Have you tried communicating using the ver 2.0 driver? There are a couple of reasons why the "interfaces sequentially verified" might fail, like if an instrument is cabled to the card. The best way to check communication is to try IBIC.
Open IBIC:
Measurement and Automation explorer >> Tools >> NI-488.2 >> Interactive Control
Go to this webpage [broken link removed]
Below the screen shot is a list of commands with definitions of the commands and parameters. Type the commands into the IBIC window.
When you write a command to the instrument, try using a simple query that the instrument will respond to. The example used in the example is *IDN?, but that is part of the 488.2 standard and may not work for your instrument
. Look in your instrument's user manual for an appropriate command.
If the instrument does not respond with IBIC, then there may be something wrong with the operating system.
If IBIC does not work, submit your question via email at www.ni.com/ask. Let me know which step failed and include a NI-SPY capture of the IBIC failing and a spy capture of the Getting Start Wizard failing. Put my name at the top of the email so I will know it's for me.
NI Spy Tutorial
Ray K
NI Applications Engineer