05-24-2013 04:16 AM
Hai..
This is Rajavel from Finland. We have Keithley 236 and GBIP cable and also labview 2012 vrsion software.
I first installed Lab view then i installed NI max (ni4882_302) verson. Then i download the Keithly 236 drivers.
After installing NII max i go device and system then i double click GPIB+HS 'GPIB0' and the press scan for the instruments and i got GPIB0::14::INSTR. In that i got the IP address (14) and select Labview 2012 version then i click button go.
It go to the labview program now is shows instrument driver finder configure search in that i select the conected instrument and i will show the code of my connected instrument which i got it NI max.if i go that way means i didnt get results in that .
After that i go the installed drivers in that it showed agilent, Keithly 2000, Keithly 24xx and Keithly 27xx after that if i select any one from keithly it shows some simple VI programs but i doestn communicate with my instrument.
So can you solve this issues for me..
Regards
05-24-2013 08:53 AM
the Keithley 236 driver is available here
This "Legacy" GPIB instrument predated the IEEE 488 standard. It does not support *IDN? so MAX will not be able to detect what the model number of the device is and the Driver Finder won't be able to search properly.
Consider using a modern instrument.