Hi,

can anyone help me on the issue I met for GPIB control? The romote controlled instrument is LeCroy waverunner LT584M. I programmed the code ibwrt(scope, "C1:VDIV 2V", 10); to try to set scope volt div as 2v/div. however the scope was set as 20V/Div.

whatever what value I pass to scope to set volt div, the scope always converts the value 10 times larger. While when I try TDIV any other setting commands, so such problem happened.

 

Anyone know why?

 

Thanks,

CC

 

0 Kudos
Message 1 of 4
(3,606 Views)

@uranus wrote:

Hi,

can anyone help me on the issue I met for GPIB control? The romote controlled instrument is LeCroy waverunner LT584M. I programmed the code ibwrt(scope, "C1:VDIV 2V", 10); to try to set scope volt div as 2v/div. however the scope was set as 20V/Div.

whatever what value I pass to scope to set volt div, the scope always converts the value 10 times larger. While when I try TDIV any other setting commands, so such problem happened.

 

Anyone know why?

 

Thanks,

CC

 


What does the programmer's manual say?

 

If indeed this is a bug with the instrument's firmware, you need to contact LeCroy directly.

 

 

 

 

 

0 Kudos
Message 2 of 4
(3,588 Views)

It almost sounds like you set the probe to be a 10X.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 3 of 4
(3,584 Views)

@crossrulz wrote:

It almost sounds like you set the probe to be a 10X.


It sounds EXACTLY like that.  The V/Div setting on those scopes does not adjust for probe gain.


"Should be" isn't "Is" -Jay