GPIB-USB-HS won't install code 39 error

Solved!
Go to solution
Solution
Accepted by topic author AndyV

Hello Kurt and Matt,

After a legnthy process, here is what I found-

Prior to installing the GPIB-USB-HS drivers, I had installed some Hewlett-Packard/Agilent IVI/VISA Drivers for another instrument. I found that I had to uninstall these drivers to get rid of the code 39 errors. Apparently the Agilent drivers (of the same name "visa23.dll") will check to see if they are still OK. when they detect they have been overloaded, they rewrite themselves from cache I think. After I completely unistalled the Agilent VISA software, I was able to avoid the code 39 error.

I have pasted in a partial transcript of the solution thread with hopes that helps or at least points you in the right direction(starts at bottom but solution is last couple of e-mails):

Ben,

BTW, I had one other machine that had that HP package on it and that

would give me some problems too on installation of one of my apps that

made use of the VISA packages. When I would start up the app, it would

tell me that it couldn't find visa32.dll but of course it was right

there in the system32 directory. Maybe since the machine I was

developing on had the problem with the conflicting software and I would

build the app on that machine that it would pull the wrong DLLs and put

them in the setup and install package. Alternatively, it could be like

you said too, that after installation, the Agilent package would

overwrite the NI DLL. I guess I can't have the two packages on the same

machine at once. Also, as a help for future debugging, when I was

removing the Agilent packages from the other machine I saw the name of

the one remaining one that I had to remove before the NI-PAL would

start. It was called the "IVI Shared Components" in the windows removal

list. If you get another code 39 problem tell them to look for that one

too. The Agilent 34970A install puts at least 3 different items in the

uninstall list and that's one of them.

Best regards,

Andy

-----Original Message-----

Note: Your reference number is included in the subject field of this

message. It is very important not to remove or modify this reference

number, or your message may be returned to you.

Andy,

You are welcome for the support. I offered it to you becasue I had seen

the effort that had gone into the GPIB error 39 in the past without a

solution. In the future, Agilent IVI software will be something we look

for with installation issues involving these devices. Thanks for

providing the detailed information about your Agilent software. We use

these kinds of notes from past service requests very often.

Regards,

Benjamin

Applications Engineer

National Instruments

http://www.ni.com/support

------------------------------------------------------------------------

------

Thank you for your unsolicited support. I hope this knowledge can help

to solve at least the one other individual that had the problem. And

maybe more if the right questions get asked. To review the history, I

was using an rs-232 connection to an Agilent 34970A and had loaded their

software (agt34970A_XL.exe) to capture some temperature and voltage

data. That had the IviSharedComponents1.3.0.msi installer as part of it

apparently. It installs as several parts that show up in different

places on the windows uninstall list, so when I uninstalled it,

apparently I missed that piece. I notice that there are still dll pieces

in the system32 directory though. Do you think they may conflict? One

that I notice is "visa32.Agilent Technologies.dll". When I examine the

file properties, it's "product version" is 14.2.8931.1 but the "file

version" is Version:78.2.8931.1. The "internal name" and "original file

name" is visa32.dll. If anybody uses those properties to identify the

file, I'm sure that would screw things up. The other piece that's

obvious is "visaConfMgr.dll" (visa conflict manager) from the IVI

foundation but that name is less conspicuous. I'm not sure what product

installed that one.

On the persistence (stubbornness), Yes I admit to it. I've been working

on computers since the 70's- the first I built was a intel 8008-1 at a

screaming 800 kilohertz. They are all the same - no matter how smart you

allocate the memory or try to hide the software innards or how many

cpu's and instruction que's you have. It just gets more messy, faster.

Regards,

Andy

-----Original Message-----

Note: Your reference number is included in the subject field of this

message. It is very important not to remove or modify this reference

number, or your message may be returned to you.

Andy,

That is awesome. I would agree with you that the problem was the

Agilent IVI drivers. They use common components with some of our

software (GPIB,

VISA) and could easily cause a conflict. In the past, we have seen

problems with software called Agilent I/O Libraries v14.1. That

software runs a watchdog service that constantly monitors some dll's.

If we install our driver and it overwrites one of the common components,

their service immediately replaces it with theirs. They might be doing

something similar with IVI now. I have to say I'm a bit surprised that

you figured this out

- it's not often that I see this level of effort on the part of

customers.

It's great that the GPIB-USB-HS now works in your computer.

If all NI processes are no able to start correctly, there is no need to

perform the steps I suggested. Those steps were only to help start the

NI-PAL Service Manager (nipalsm) process. Thank you for continued and

concentrated effort on this issue.

Best regards,

Benjamin

Applications Engineer

National Instruments

http://www.ni.com/support

------------------------------------------------------------------------

------

Ben- Good news! After clearing the machine with msiblast then installing

my downloaded copy of ni488.2 v2.7 as I indicated that I would do in my

previous e-mail, I am now able to insert the GPIB-USB-HS, the driver

loads and MAX communicates with a device I have connected to the

interface using VISA. Eureka! And I didn't have to wipe my disk!!! Yea!

Ok-I can still try your suggestion and I will but now I've got work to

do. Although I'm not sure that I can tell if it will do anything. My

guess is that the IVI library was the culprit at this point. What say

you?

Thank you for your support,

Andy

-----Original Message-----

Note: Your reference number is included in the subject field of this

message. It is very important not to remove or modify this reference

number, or your message may be returned to you.

Andy,

I have an idea that might help in allow nipalsm to run. By the way,

nipalsm is the NI-PAL Service Manager and it is required as it helps

load other programs.

1. Close all NI programs such as MAX, LabVIEW etc.

2. Browse to C:\Documents and Settings\All Users\Application

Data\National Instruments\MAX\Data (you will need to show hidden files

and folders to see the Application Data folder) and clear all the

contents of this directory.

Sometimes, you will not be able to clear all the contents of this

directory. In this case, open Task Manager and end all NI related

process.

Then, try clearing the contents of Data directory.

Tip: All NI related processes in Task Manager start with nixxxx, as

shown in the image below.

3. Restart your computer.

After you do this, see if you can get the NI-488.2 Configuration Manager

to load from services.msc.

Regards,

Benjamin

Applications Engineer

National Instruments

Message Edited by Support on 10-20-2009 05:15 PM