HP 4155A High voltage detected error
Hello.
I have a question about the CAUTION in HP 4155A instrument.
We have a problem to control the instrument because of the error in the picture below.
Before the error occurs we used the instrument well by GPIB0 communication.
(2,086 Views)
No error attached.
Can you perform the same task manually?
I suspect you have not setup the interlock properly.  See - https://forums.ni.com/t5/Instrument-Control-GPIB-Serial/HP4155A-Fixture-Open/m-p/4056371#M81533
Craig
(2,073 Views)
Oops, thank you for your comment.
I attach the picture of the error.
(2,066 Views)
Accepted by topic author JJ.Lee
Yikes, that's not an operation error but an SMU error from the calibration/diagnosis page.  I expect it appears even if you cycle the power?
Looks like the SMU has been damaged and needs repair.  You should contact Keysight for support.
Good luck.
Craig
(2,025 Views)
Yes, it appeared right after when I turn on the power.
Thank you for your help.
I'll contact  Keysight for support.
Thank you so much.
Jea Jung Lee.
(2,006 Views)