First, you don't have to have two separate sessions for Write and Read. Some session attributes are "Local Attributes" that are managed per session, thereby you may have to set the same attributes twice for two sessions for each. Instead, having only one session simplies them.
Here's a VISA-COM C# example that just queries *IDN? response to an SCPI instrument, with 19200/N/8/1 and X-Flow. (You may change serial communication conditions according to your actual instrument to connect.)
IResourceManager rm = null;
IMessage session = null;
string strRd;
System.Array block;
try
{
rm = new ResourceManager();
session = (IMessage)rm.Open("ASRL3::INSTR", AccessMode.NO_LOCK, 0, "");
ISerial ser = (ISerial)session;
ser.BaudRate = 19200;
ser.DataBits = 8;
ser.StopBits = SerialStopBits.ASRL_STOP_ONE;
ser.Parity = SerialParity.ASRL_PAR_NONE;
ser.FlowControl = SerialFlowControl.ASRL_FLOW_XON_XOFF;
// Read until LF(0x0A) byte for *IDN? response
session.WriteString( "*IDN?\n");
strRd = session.ReadString( 100);
// Read exactly 8 bytes for *IDN? response
// (with waiting for 8 bytes get available for read)
session.TerminationCharacterEnabled = false;
session.WriteString( "*IDN?\n");
while( ser.BytesAvailable < 😎
{
;
}
block = session.Read( 8);
}
catch( System.Runtime.InteropServices.COMException ex)
{
MessageBox.Show( ex.Message, "error 0x" + Convert.ToString(ex.ErrorCode, 16));
}
finally
{
session.Close();
}
In the TRY block, there are two different approaches for reading, 1)typical ASCII read until LF(0x0A) terminate, 2)block read for specific byte length. As for typical ASCII command IO bases, the 1st approcah is much easier. As for arbitrary block read, you should disable the TerminationCharacterEnabled property to avoid unexpected termination for the case when an LF code was encountered. Also you can check the byte length that is available in the read-buffer by using ISerial.BytesAvailable property before reading.
Hope this helps.
Makoto Kondo
01-11-2006 09:23 PM
01-11-2006 10:10 PM
01-11-2006 11:07 PM
I rand the program again, and got an exception when reading the response to the "IDN*"command: "Exception from HRESULT: 0x80040072". Do you know what it means?? Maybe I have the session configured wrong? I only used the same parameters that you had shown.
I have also incorportared the example code you gave me into my program, and I am still getting the same [incorrect] data when I communicate with my device. For example, when I send a command to the device and read the response, the scope reads 8 bytes coming back in the response, but the VISA ReadString function only gives me 5 bytes. And sometimes I get these characters with strange values that do not even map into ASCII characters, and when I convert the values to decimal values I get very high numbers, like 8322. How can that be a single byte?
Here's some sample output from my program:
'DefaultDomain': Loaded 'c:\windows\microsoft.net\framework\v1.1.4322\mscorlib.dll', No symbols loaded.
'E2C_BI': Loaded 'C:\Software\E2C BI source\E2C_BI\bin\Debug\E2C_BI.exe', Symbols loaded.
'E2C_BI.exe': Loaded 'c:\windows\assembly\gac\system.windows.forms\1.0.5000.0__b77a5c561934e089\system.windows.forms.dll', No symbols loaded.
'E2C_BI.exe': Loaded 'c:\windows\assembly\gac\system\1.0.5000.0__b77a5c561934e089\system.dll', No symbols loaded.
'E2C_BI.exe': Loaded 'c:\software\e2c bi source\e2c_bi\bin\debug\interop.visacomlib.dll', No symbols loaded.
Exception from HRESULT: 0x80040015.
error 0x80040015
€
Exception from HRESULT: 0x80040015.
error 0x80040015
8222
Exception from HRESULT: 0x80040015.
error 0x80040015
402 70 2 74 1 1
Exception from HRESULT: 0x80040015.
error 0x80040015
8364
Exception from HRESULT: 0x80040015.
error 0x80040015
8364
01-11-2006 11:43 PM
The error code 0x80040072 is probably from VISA-COM as it defines -- E_VISA_RSRC_BUSY (The resource is valid, but VISA cannot currently access it). I have no experience to encounter this error code, however, some of hardware resource is conflicting by other VISA sessions (if any same or similar resource is being occupied). That error is very VISA implementation specific and hardwware IO specific. I recommend you to check if any other program has an access the same RS-485 port at the same time.
As for strange high byte characters, I believe ANSI-UNICODE conversion is concerning. VISA-COM treats every STRING parameter and return as a UNICODE string. This is a generic COM rule defined by Microsoft. Therefore ReadString() method that returns BSTR (= W_CHAR* type) type internally converts ASCII based GPIB/Serial response string to UNICODE. If the instrument response includes only ASCII characters below 0x7F, the ReadString() returns the same image without problem. However, if the response includes a high-byte char, ANSI-UNICODE conversion misunderstands as if the given high-byte was a multi-byte portion of Asian SHIFT-JIS(Japan), GB2312(China), BIG-5(Taiwan), or KS(Korea) encodes Therefore, if your instrument may return a response including one or more high-bytes, you should use Read() method rather than ReadString(). The VISA-COM IMessage::Read() method returns a BYTE array (strictly a COM-defined SAFEARRAY of unsigned char type) that has the exactly same contents as the instrument returned. No UNICODE conversion is applied because the return type is BYTE array (not STRING !). Through the .NET COM interop layer, the Read() method returns .NET System.Array type that will actually contain a BYTE array.
As for unexpected termination, VISA's TerminationCharacterEnabled property may be activated. The default of this is TRUE and the defalt TerminationCharacter applied is 0x0A. Setting TerminationCharacterEnabled = false will avoid unexpected termination even if LF (0x0A) was sent from the instrument.
01-12-2006 09:41 AM
Hi Goharik,
I noticed from your posts here that you're using the VISA-COM API through interop in C#. National Instruments provides a full-featured native .NET API that ships with NI-VISA. If you have Measurement Studio installed, it will be automatically installed when you install NI-VISA. If you do not have Measurement Studio, you might need to explicitly enable installation of the .NET API. It should appear in the feature tree under the "Development Support" as ".NET Framework 1.1 Languages Support". This feature installs a set of .NET assemblies into the GAC, as well as into the "C:\Program Files\National Instruments\MeasurementStudioVS2003\DotNET\Assemblies\Current" directory. If you have Measurement Studio installed, you can use the Add/Remove Class Library Wizard to add references to the NI-VISA assemblies to your project. You might find that this .NET API works better than using the IVI-COM components through interop.