How to perform serial spolling by VISA?

I'm trying to controll KEITHLEY 6517 by Excel Macro with VISA. But, the function of "spoll" has not be expressed by VISA. I'm not good at system term in English so I cannot understand what manual says. Please give me the solution how to perform serial spolling by VISA. The copy of a part of KEITHLEY manual which referes to "serail spolling" is attached.
0 Kudos
Message 1 of 2
(3,410 Views)
Naohito:

The C function call you want is viReadSTB, or in LabVIEW, it is VISA Read Status Byte. Reading the status byte over GPIB performs an IEEE-488.1 serial poll.

Dan Mondrik
Senior Software Engineer, NI-VISA