I need help setting up GPIB with Hitachi F-4500 spectro fluorometer

Hello

 

I recently purchased a used Hitachi F-4500 spectro fluorometer and installed the necessary GPIB drivers on my windows 11 PC and made all changes in order to communicate with the device.

But the application (FL solution) freezes when I start it.

I recorded the instruction flow (see attachment) and notice that the last instruction is IBWAIT(GPIB0,0x1000). What could cause this problem?

 

Any help is appreciated.

Best regards

Juergen

 

 

 

Message 1 of 4
(1,091 Views)

Have you reached out to the manufacturer for help/support?

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 2 of 4
(1,051 Views)

Hello Juergen!

 

One of the old PCs (Windows 2000) to which our F-4500 was connected via GPIB broke down recently (Institute of Biophysics, JKU Linz). I have been trying to get it going again on another old PC (Windows XP). Yet, after several attempts (also drawing inspiration from your other posts in this forum) I have ended up with the very same sequence of commands (using NI Spy) that you show when using the Hitachi Software to search for the instrument. To that effect, I was wondering whether you could resolve the issue back then, and if yes, if you could hint towards a possible solution for us.

 

Thank you kindly and best regards!

(also) Juergen

0 Kudos
Message 3 of 4
(672 Views)

Hello!

 

Eventually, I got the F-4500 to run with an old Windows XP PC and a Windows 11 laptop running a Virtual Machine with Windows 2000. Below is a short description of how I did that:

 

Windows 11, Virtual Machine running Windows 2000

The operation is possible in a Virtual Machine running Windows 2000 on my laptop (HP EliteBook 830 G5 running Windows 11 Education Version 10.0.22631 Build 22631). To install a VM running Windows 2000 in Oracle VM VirtualBox, I loosely followed the instructions given herein: https://i12bretro.github.io/tutorials/0071.html. Notably, the installation of Windows 2000 in the Virtual Machine worked (surprisingly) smoothly!

 

In Windows 2000 (VM), several programs were installed:

  • FL Solutions: I installed the program from four executables which I had on a CD-ROM (those were probably copied there from several diskettes which we own as well).
  • DAEMON Tools Lite: This program was required for installing the NI-488.2 2.4.0 ISO file.
  • NI-488.2 2.4.0: Note that, albeit not tested thoroughly, version 2.5 did not work right away.

For convenient installation, access to a folder on the PC running the VM by Windows 2000 in the VM was obtained by VM menu bar: Devices: Insert Guest Additions CD Image (the folder eventually appeared as “Hitachi on ‘VBoxSvr’ (Z:)” in My Computer (Win+E) on Windows 2000 (VM)).

 

Following successful installation of FL Solutions and NI-488.2 drivers, the GPIB connection was configured. Therefore, I followed the instructions in the “User’s Instruction Manual for F-2500/F-4500 Fluorescence Spectrophotometer”)! That said, the procedure slightly deviated from the illustrated description in the manual. First, I used a GPIB-USB-HS adapter, not the PCI-GPIB card. Second, the driver version (NI-488.2 2.4.0) I used may have been different to the one described by Hitachi (version not specified). Third, the installation description may have applied to Windows XP rather than Windows 2000. In particular, the necessary programs are not accessible via the Device Manager in Windows 2000 (neither are they in Windows XP running NI-488.2 2.4.0 though, again indicating that Hitachi describes a different version of NI-488.2). That said, one can use NI Measurement & Automation to configure the GPIB connection:

  • Change “I/O Timeout” to 3 sec
  • Release the check mark at “Automatic Serial Polling” and select “Assert REN when SC”

The interface properties can be changed using the program GPIBConf that can be found in the installation folder of the National Instruments software. In the specific case of my installation in the VM, this is under “C:\Program Files\National Instruments\NI-488.2\Bin”.

  • Select “DEV1” of “Device Name” and change DEV1 to FL. Change “I/O” of “Timeouts” to 3 sec and “Serial Poll” also to 3 sec. Select “Readdress” and then click OK.

 

Following similar instructions, the fluorometer was set up to work with an old Windows XP PC

Notably, the Windows XP PC is connected to the fluorometer using a PCI-GPIB card (perhaps the card shipped with the Hitachi fluorometer). Otherwise, the setup is practically identical.

 

I hope this helps someone!

Kind regards!

 

Juergen Pfeffermann