I recently purchased a used Hitachi F-4500 spectro fluorometer and installed the necessary GPIB drivers on my windows 11 PC and made all changes in order to communicate with the device.
But the application (FL solution) freezes when I start it.
I recorded the instruction flow (see attachment) and notice that the last instruction is IBWAIT(GPIB0,0x1000). What could cause this problem?
Any help is appreciated.
Best regards
Juergen
05-14-2023 01:44 AM
Have you reached out to the manufacturer for help/support?
02-23-2024 07:24 AM
Hello Juergen!
One of the old PCs (Windows 2000) to which our F-4500 was connected via GPIB broke down recently (Institute of Biophysics, JKU Linz). I have been trying to get it going again on another old PC (Windows XP). Yet, after several attempts (also drawing inspiration from your other posts in this forum) I have ended up with the very same sequence of commands (using NI Spy) that you show when using the Hitachi Software to search for the instrument. To that effect, I was wondering whether you could resolve the issue back then, and if yes, if you could hint towards a possible solution for us.
Thank you kindly and best regards!
(also) Juergen
07-26-2024 09:00 AM
Hello!
Eventually, I got the F-4500 to run with an old Windows XP PC and a Windows 11 laptop running a Virtual Machine with Windows 2000. Below is a short description of how I did that:
Windows 11, Virtual Machine running Windows 2000
The operation is possible in a Virtual Machine running Windows 2000 on my laptop (HP EliteBook 830 G5 running Windows 11 Education Version 10.0.22631 Build 22631). To install a VM running Windows 2000 in Oracle VM VirtualBox, I loosely followed the instructions given herein: https://i12bretro.github.io/tutorials/0071.html. Notably, the installation of Windows 2000 in the Virtual Machine worked (surprisingly) smoothly!
In Windows 2000 (VM), several programs were installed:
For convenient installation, access to a folder on the PC running the VM by Windows 2000 in the VM was obtained by VM menu bar: Devices: Insert Guest Additions CD Image (the folder eventually appeared as “Hitachi on ‘VBoxSvr’ (Z:)” in My Computer (Win+E) on Windows 2000 (VM)).
Following successful installation of FL Solutions and NI-488.2 drivers, the GPIB connection was configured. Therefore, I followed the instructions in the “User’s Instruction Manual for F-2500/F-4500 Fluorescence Spectrophotometer”)! That said, the procedure slightly deviated from the illustrated description in the manual. First, I used a GPIB-USB-HS adapter, not the PCI-GPIB card. Second, the driver version (NI-488.2 2.4.0) I used may have been different to the one described by Hitachi (version not specified). Third, the installation description may have applied to Windows XP rather than Windows 2000. In particular, the necessary programs are not accessible via the Device Manager in Windows 2000 (neither are they in Windows XP running NI-488.2 2.4.0 though, again indicating that Hitachi describes a different version of NI-488.2). That said, one can use NI Measurement & Automation to configure the GPIB connection:
The interface properties can be changed using the program GPIBConf that can be found in the installation folder of the National Instruments software. In the specific case of my installation in the VM, this is under “C:\Program Files\National Instruments\NI-488.2\Bin”.
Following similar instructions, the fluorometer was set up to work with an old Windows XP PC
Notably, the Windows XP PC is connected to the fluorometer using a PCI-GPIB card (perhaps the card shipped with the Hitachi fluorometer). Otherwise, the setup is practically identical.
I hope this helps someone!