04-03-2006 10:31 AM
04-04-2006 06:20 AM
are baudrate etc correct
did you try with hyperterminal and got an answer than sure the lines are ok
did you send crlf after the *IDN?
so the instrument sees the End Of Message that is standard in gpib but must be sent in rs232
Try to swap the wires on pin 2 and 3
we call rs232 the Regular Soldering 2->3->2 standard
are baudrate number of bits, etc correct
and sometimes the instrument looks on DTS or RTS to see if the other side is powered up.
success
04-04-2006 06:21 AM
Hi,
what kind of instrument are you communicating to? Have you already cheked on the instrument programming manual that it accept that command? I mean there are some instruments that use for istance "ID?" or "IDN?" or anything else instead of "*IDN?".