Inficon Quadrex 200 with RS232 coomunication

Hello everyone
 
I was trying to communicate computer with Inficon Quadrex 200(mass spectrometer). but I got some error message.
 
Can you see program and give me some advices if you know the problem in my program.....
 
Thanks...
Download All
0 Kudos
Message 1 of 4
(3,735 Views)
That's not an error, it's a warning. When you get an actual error, the status Boolean turns to a red 'x'. The warning is normal when you disable termination character detection because the default behavior is to have it enabled and then VISA Read will stop its reading when the termination character is detected. It's just alerting you to the possiblity that there may be bytes left in the srial buffer.
0 Kudos
Message 2 of 4
(3,722 Views)

Thaks your help

 

I will test it again...

0 Kudos
Message 3 of 4