Hello,
Are you using an instrument driver for your Agilent device? If so, is this one you downloaded from the National Instruments Device Network? If you so, there may be un updated version of that driver available now. Also, what do you mean by the SRQ is not generated in the GPIB software? Is it possible that the device is asserting it and then deasserting it once the controller responds to the service request? In other words, how do you know SRQ is not generated?
Let me know and I will look into the issue further. Have a great day!
12-31-2006 11:58 PM
01-02-2007 04:46 PM
01-04-2007 01:07 AM
01-05-2007 02:54 PM
02-07-2007 06:07 PM
03-12-2007 11:48 AM
Hi again,
Sorry for the long delay in getting back to you. Version 1.7 should work, but 2.4 does not. Make sure you are setting the ibcAUTOPOLL to 1 using ibconfig. Doing this should fix the problem. Try this out and let me know how it works.
Have a great day!
07-26-2007 04:19 PM