Keithley GPIB only works on Ch 1?

The GPIB settings are rather simple - on/off and pick a channel. It is on and responds to *IDN? 

 

When set to channel 1, a GPIB scan over the LAN shows all the instruments. If I pick a different channel (different from all the other GPIB devices of course), it does not show up on the scan with the VISA interactive control like the others. I have disconnected all other instruments as well and swtiched with known working cables. 

 

....as I was typing I noticed something. In MAX, I enter the LAN Device Name: gpib,0,8 and press validate. 

MAX 09-22-2015 2700-b.png

 

 

 

There it is.....

 

VIC 09-22-2015 2700-b.png

 

 

 

Then, the refresh on the VISA interactive control shows the Keithley on GPIB channel 8 through the TCPIP interface. Not sure why I have to manually type that in the MAX. Good news is that it is connected, the bad news is that I don't know why. The other instruments don't need any special help. Should I worry any further? If I change the channels or add other instruments, they seems to be found automatically - the Keithley needs a little bump I guess. 

 

 

 

 

0 Kudos
Message 11 of 12
(2,198 Views)
It's address and not channel. As long as the new address is visible in MAX after doing a refresh, you should be just fine.
0 Kudos
Message 12 of 12