Keysight 33522B, 2000 Series Keysight Scope Self Test Failure
Hello Everyone,
I am working on Scope, Waveform Generator, and Amplifier. I am using DLL which is provided by Manufacturer.
My problem is in Self Test and All of my devices have the same problem. While self-test working, the device returns false request but ı see on the screen the self-test continue and it pass after 5-10 second.
How can ı do solve my problem? It looks like a timeout problem I think. I know VI_ATTR_TMO_VALUE attribute is about timeout attribute. But ı couldn't find Attribute ID.
(3,542 Views)
Seeing how this instrument does accept SCPI commands, try to send the following:
*OPC?\n
To verify that the system is not currently waiting on any other commands.  It should respond with a "1" then send the
*TST?\n
Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
(3,523 Views)
*OPC?\n
*TST?\n
When .ı send command on NI MAX, *OPC?\n returns 1, and *TST?\n returns 0. Everthing is normal. I am working  on TestStand. Self Test function returns respond before the finish self test function.
(3,504 Views)
I saw a note in code. That is  "* Please set VI_ATTR_TMO_VALUE to 30000 ms or greater." How can ı change  VI_ATTR_TMO_VALUE attribute.
(3,500 Views)
Accepted by topic author standalone
I solved this problem.
viSetAttribute (io, VI_ATTR_TMO_VALUE, 30000);
 I added this code to my device source code.
(3,499 Views)
