Hi all, I'm working on a program to read and write to a piece of equipment using GPIB.
At the moment NI Max does not pic up that the device is connected through GPIB but the labview program does.
Origionally I thought that the computer was not regestering the GPIB PCI Card but when I went into device manager it showed that the computer was picking up on the card.
Is there a reason or setting at to why max does not pick up the GPIB device and the labview program does??
The meter is GPIB0::11::INSTR
Please find images attached, and any advice would be helpful.
Thank you.
T.J.
11-07-2013 08:56 AM
MAX uses the low level functions from NI-488. LabVIEW populates the IO control drop down from VISA calls.
That ADLINK device does not play nice with NI 488 although it's VISA implementation does follow VISA conventions.
(Buy cheap get cheap)
11-07-2013 09:10 AM
Hi Jeff, thank you so much for replying.
So what you are saying is its the actual PCI card itself?
Thank you.
T.J.
11-07-2013 09:15 AM
11-14-2013 10:31 PM
Try: MAX-> Tools -> NI VISA-> VISA Options, you'll see the configuration pannel, then General Settings -> pass-ports, you'll see a List of Pass-ports, make sure tick up all checkboxes. (GPIB Pass-port is un-checked defaultly.). Then save & exit, relaunch MAX, you might be able to see your third party GPIB card.
It works for Agilent any way.
Good luck.