Missing byte in GPIB transfer
in my measurement configuration I have a network analyzer connected to a PC using GPIB interface (NI PCI-GPIB, default setting).
The control software is HTBasic from TAMS which always uses the GPIB in the asynchronous way.
The controller should be fast enough (dual core @3 GHz)
Sometimes the transfer of binary data between NWA and PC fails. I learned that one byte will be lost.
I tried to find the cause of this, but in vain up to now.
More details:
A GPIB analyzer protocol displays all bytes, in the parallel NI-Spy log one byte is missing.
Here is the GPIB analyzer protocol, the relevant byte is underlined:
          --  --  --        a   00001010  0 0 0 1 0  0 1 1  DAB
          --  --  --        1d  00011101  0 0 0 1 0  0 1 1  DAB
          --  --  --     ?  3f  00111111  0 0 0 1 0  0 1 1  DAB
          --  --  --        e2  11100010  0 0 0 1 0  0 1 1  DAB
          --  --  --        90  10010000  0 0 0 1 0  0 1 1  DAB
          --  --  --        91  10010001  0 0 0 1 0  0 1 1  DAB
          --  --  --     >  3e  00111110  0 0 0 1 0  0 1 1  DAB
          --  --  --        a7  10100111  0 0 0 1 0  0 1 1  DAB
          --  --  --        e0  11100000  0 0 0 1 0  0 1 1  DAB
          --  --  --        1c  00011100  0 0 0 1 0  0 1 1  DAB
The export of NI-Spy in this section shows the following result (again with underline in the important section):
1227750.  Asynchrone I/O neu synchronisiert für ibrda()
Prozess-ID: 0x00000900         Thread-ID: 0x0000090C
Startzeitpunkt: 09:40:40.809       Aufrufdauer 00:00:00.000
ibsta: 0x2164       iberr: 0             ibcntl: 457(0x1c9)
Ausgangspuffer
00000000:  8F 77 7A BF BB 42 53 3E 95 23 79 BF C7 3A 6B 3E  .wz..BS>.#y..:k>
.etc
000001C0:  59 FE 1C 3F ED FE 91 3E 0A                       Y..?...>.
1227751.  ThreadIbcntl()
1227752.  ibrda(GPIB0, 0x01CFC503, 5159 (0x1427))
1227753.  ibwait(GPIB0, 0x0000)
1227754.  Asynchrone I/O neu synchronisiert für ibrda()
ibsta: 0x2164       iberr: 0             ibcntl: 17(0x11)
Ausgangspuffer
00000000:  0A 1D 3F E2 90 91 3E A7 E0 1C 3F E0 3E 91 3E 8F  ..?...>...?.>.>.
00000010:  D3                                               .
1227755.  ThreadIbcntl()
What is obvious is that
- the binary stream is split up at bytes 0x0a (although the application reads in the stream with one command: "ENTER @Stream;Array(*)") which doesn't make a problem.
- some bytes before the missing byte there are two consecutive bytes 0x0a (may be by chance)
- after the missing byte the stream is split up again, but here (there is only one occurrence of this) NOT at 0x0a
What I have tried:
- It does not help setting the interface to "No autopolling"
- No other device is on the bus
- The NWA does not send a SRQ
What I want to know:
- can the application have an influence on the content of the NI-Spy? Is it the driver or the application?
- How can I prevent this?
I hope there is someone being able to help as I don't know how to proceed.....
Many thanks in advance.
(4,302 Views)
after inspecting the two logs for several times I was not able to determine where I should find a lost byte.
Can you specify where there is something lost?
Best regards and a merry christmas,
Frank
(4,260 Views)
0x0A is the default termination character.  so with term char enabled the read stops when it sees a 0x0A
Drop one of these before your read and read the number of bytes you expect
(4,239 Views)
Dear Frank,
thank you for checking my request. Merry Xmas as well. This is the reason why the reply is so late...
You are right, I copied the wrong section and marked the wrong 0x1c byte. The relevant logs are shortly afterwards:
 GPIB-Analyzer:
          --  --  --        8f  10001111  0 0 0 1 0  0 1 1  DAB
          --  --  --        d3  11010011  0 0 0 1 0  0 1 1  DAB
          --  --  --        c7  11000111  0 0 0 1 0  0 1 1  DAB
and the NI-Spy log is here:
Ausgangspuffer
00000010:  D3                                               .
1227756.  ibrda(GPIB0, 0x01CFC514, 5142 (0x1416))
ibsta: 0x0       iberr: 0             ibcntl: 0(0x0)
1227757.  ibwait(GPIB0, 0x0000)
ibsta: 0x64       iberr: 0             ibcntl: 0(0x0)
1227758.  ibwait(GPIB0, 0x0000)
1227759.  ibwait(GPIB0, 0x0000)
1227760.  ibwait(GPIB0, 0x0000)
1227761.  ibwait(GPIB0, 0x0000)
1227762.  ibwait(GPIB0, 0x0000)
ibsta: 0x2164       iberr: 0             ibcntl: 626(0x272)
1227763.  Asynchrone I/O neu synchronisiert für ibrda()
Ausgangspuffer
00000000:  3F C7 14 91 3E D3 2A 1D 3F 8F 61 90 3E 5A DC 1C  ?...>.*.?.a.>Z..
I ask you to look again into this code...
Sorry for the inconvenience caused so far.
Best regards
Andy
(4,208 Views)