No Listeners on Bus
i am trying to get a simple applicaiton to write to a gpib device. the interface is a GPIB-ENET device. i open the interface using ibdev, clear the device using ibclr, check if the device exists using ibln and then write the string " E " to the device (a 2100 computotor indexer). all the commands except the ibwrt return a 0x100 status and no error. the ibwrt returns an ENOL error.
i have tried this same command with the gpib interactive interface gpibintctrl using only ibdev and ibwrt and the command suceeds.
can someone help me witrh what i am doing wrong please.
i have attached the source code.
thanks in advance
regards
sean
#include
#include
#include
#include "ni488.h"
#define ARRAYSIZE 1024 // Size of read buffer
#define ESCAPE 0x1b
// Response byte after serial polling device.
#define SERIALPOLLRESPONSE 0x60
int Dev[8][31];
char ValueStr[8][31][ARRAYSIZE + 1];
char ErrorMnemonic[21][5] = {"EDVR", "ECIC", "ENOL", "EADR", "EARG",
"ESAC", "EABO", "ENEB", "EDMA", "",
"EOIP", "ECAP", "EFSO", "", "EBUS",
"ESTB", "ESRQ", "", "", "", "ETAB"};
char errmsg[1024], str[1024];
int device, ud, board;
long count;
void main(int argc, char *argv[]) {
short exists;
device = 3;
ud = ibdev(board,device,0,13,1,0);
sprintf(errmsg,"Opening GPIB device on board %d at primary address %d. Error = %d. Status = 0x%x\n", 0, device, iberr, ibsta);
printf("%s",errmsg);
ibclr(ud);
sprintf(errmsg,"Clearing GPIB device on board %d at primary address %d. Error = %d. Status = 0x%x\n", 0, device, iberr, ibsta);
ibln(board,device,0,&exists);
sprintf(errmsg,"Checking GPIB device on board %d at primary address %d. Error = %d. Status = 0x%x\n", 0, device, iberr, ibsta);
printf("Device at adress %d exists status = %d\n",device,exists);
strcpy(&str[0]," E ");
count = strlen(str);
ibwrt(ud,&str[0],count);
sprintf(errmsg,"Writing to GPIB device %d on board %d at address %d, %d number of characters. Error = %d. Status = 0x%x.\nString : \"%s\", count = %d\n",ud,0,device, count, iberr, ibsta, str, ibcntl);
(4,559 Views)
What version of the driver are you using?
Does the device show up in MAX?
Does MAX show the device at PAD 3 as well?
Are you sure your board is the CIC?  Try sending ibsic before ibclr.
Hope this helps,
Robert Mortensen
Software Engineer
(4,546 Views)
yes, the device shows up in max. using max i can address the device and write to it. the board seems to be in cic. after stuffing around a bit, i have managed to talk to the device by inserting a sleep of one second after issueing ibdev. after that i can write toi the device without a problem. any ideas as to why?
maybe a device that requires slower time responses?
tia
(4,524 Views)
Is this an older, slower instrument?  It is very possible that it simply takes a longer amount of time to assert itself as a listener.
Robert Mortensen
Software Engineer
(4,520 Views)
yes, about at least 20 years old.  it was originally run on a PDP-11 with RSX11M+ and then on an HP700 C100 workstation running HP-UX 10.20.
(4,516 Views)
I'm not sure why pausing after the ibdev would help, but I noticed a couple of things from your code:
You don't seem to be initializing the board variable, so it could be the device handle is to the wrong board.
Also, you have mixed in a board-only ibln call with device calls, which is not recommended.
-Dan
(4,491 Views)
cocnerning the code, i noticed that and sorted it out by setting it to the appropriate value (0 in this case). i also removed all the various code that has board level calls and still require a delay. i have not experimented with how long the delay needs to be, however. i have attached the code.
#include
#include
#include
#include
#include "ni488.h"
long count;
short exists;
device = 1;
ud = ibdev(0,device,0,10,1,0);
sleep(1);
sleep(30);
strcpy(&str[0]," Z ");
(4,486 Views)