PXIe-8135 GPIB Scan for Instruments disabled
I want to connect an electronic load to the PXIe-8135. I bought a Micro D cable and connect the load . When I open the Measuremanet Explorer, I open devices and see the PCI-GPIB. The self-test was successful. But the button "Scan for instruments" is not available. It is grey and has no funktion. Has anybody a solution? The settings from the PCI-GPIB are: GPIB-ID=GPIB0; Primary Adress 0; Secondary none; Timeout 13(10s); System-Controller and Autopolling are enabled; End of write enabled; Cable HS488 0; Parallelpoll 0 (2µs); Bus-Timing 2 (500ns).
regards
Message 1 of 10
(8,701 Views)
regards
Does connecting a load automatically make the load an instrument???
Message 2 of 10
(8,692 Views)
I don't know. I think, there is an general problem maybe in the software configuration of the PXIe.
Message 3 of 10
(8,679 Views)
Hi Miks,
I've got some questions.
- What electronic load do You want to connect?
- Which GPIB-Card do You use?
- Don't You think You'll probably need a GPIB-Cable?
- Could You upload a screenshot of the MAX,
  and a MAX-Report?
Regards
Message 4 of 10
(8,626 Views)
this sounds very similar:
Solved: VISA: Access to the resource or remote machine is denied. - Discussion Forums - National Instruments
https://forums.ni.com/t5/Instrument-Control-GPIB-Serial/VISA-Access-to-the-resource-or-remote-machin...
Message 5 of 10
(8,486 Views)
please excuse the long time without an answer. The problem still exists.
I followed your link. The adress of the host PC (xxx.102) is set under "security" of the PXIe8135's NI-VISA.
But the VISA interactive control as mentioned in https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019P3vSAE&l=en-US is not working. When I select the remote controler  (xxx.230) at "Machine" an error occurs.
The error massage and Ni-MAX is attached.
The NI-Server is running on the PXIe8135 and i think the version should be compatible, because they are the same (NI-Visa 15.0.0)
Regards
Message 6 of 10
(8,316 Views)
could you check the instructions in this KB?
NI-VISA Error 0xBFFF00A7 from Real-Time Target - National Instruments
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019R76SAE&l=en-US
Message 7 of 10
(8,247 Views)
The instructions https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019R76SAE&l=en-US were checked out, but the problem still exist.
To explain better:
I want to connect a Power Control Resource (PCE Chroce 63804) to the PXIe-8135 through a GPIB HS488 0 cable.
Through MAX it´s possible to see a GPIB resource, but it´s not possible to "Scan for Instruments", as in the picture 1, so it´s not possible to follow the normal Tutorial instructions (shown in http://www.ni.com/getting-started/set-up-hardware/instrument-control/gpib-connect ).
Trying to communicate through VISA Interactive Control (version 15.0) results in Read Operation Error Hex 0xBFFF0015.
The adress of the host PC is set as written in https://forums.ni.com/t5/Instrument-Control-GPIB-Serial/VISA-Access-to-the-resource-or-remote-machin... and the KnowledgeBase. (as in picture 2)
But the VISA interactive control as mentioned in the KnowledgeBase is still not working (as in picture 3).
Best regards
Message 8 of 10
(8,110 Views)
I have exactly the same problem as Miks.
PXIe-8135 Controller, "Scan for instruments" button is grayed out, VISA interactive control throws error 0xBFFF00A7.
I have installed visa 15.5 but during the boot sequence it says "NI-VISA Server 15.0 started successfully." Could that be an indication that there is a version conflict? Re-installing VISA didn't change anything.
A solution would be nice...
By the way, USB and RS232 VISA resources work well.
Message 9 of 10
(7,733 Views)
apparently the "Scan for Instruments" Button is greyed out on RT Systems, but ti is still possible to establish a connection:
How Can I Test Communication With GPIB Instruments Connected to My Real-Time Target? - National Instruments
How Do I Control My VISA Device from Another Computer? - National Instruments
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000YIFCCA4&l=en-US
Message 10 of 10
(7,631 Views)