Our group recently got a Newport 1830-C and we ran into the exact same problem that Ryan mentioned when trying to use the "Converse with 1830-C.vi" driver. We went through the driver and figured out the problem. It's missing a 'read-done bit' polling module, so the data stream gets messed up after you send a command to the 1830-C. We wrote a polling module and added it into the driver and it works just fine now.
If anyone needs or wants to see our updated driver, I have attached the updated 1830-C library.
07-07-2005 02:03 PM
07-07-2005 02:42 PM - edited 07-07-2005 02:42 PM
Message Edited by Alwin W on 07-07-2005 03:44 PM
07-08-2005 10:16 AM
05-29-2006 05:22 PM
Is there anyone who still have problem with Newport's 1830-C Optical Power Meter?
I am using Alwin's updated driver. I have three Newport 1830-C, two of them are working fine under the new driver, but the third one is still showing the error.I can get several sets of data before it crashs. After the error shows up, if I scan instruments in Measurement and Automation, none of the devices (I have several other devices connected to this computer by GPIB) would show up. I have to unplug the GPIB to the power meter, and reconnect it. Then the Newport 1830-C would work for a while, then crash again.
Any ideas? Thank you very much
05-30-2006 06:40 PM
05-31-2006 05:19 PM
Chris,
The three 1830-C are the same, and the settings are the same. I saw the problem with the same one each time.
Yes, turning the instrument off and on again clears out the instrument.
Thank you very much.
J3000