Problem with PCI-GPIB interface card to control Agilent 8648D signal generator
Hi
I downloaded Plug and Play (project-style) driver for Agilent 8648A signal generator. I use LabVIEW 2011.
I open project file and choose COM1 from menu for VISA resource name. It should be noted that I use 82350B PCI-GPIB Interface Card.
I run the project and I get timeout error.
Is there some problem in the PCI-GPIB interface or am I missing something? I'm new to LabView and to PCI-GPIB interface, and I would really appreciate any help.
Thank you
Yasamin Gharib
(4,137 Views)
Why are you choosing COM1?
GPIB is not a com port.
(4,120 Views)
Accepted by topic author yasmain62
Since you aren't using an NI GPIB, go into MAX, select software and then NI-VISA. In the window on the right, click the VISA Options tab and then select Passports under General Settings. Check the NI-VISA Passport for Tulip.
I'm assumimg you loaded the driver from Agilent and tested communication with their software.
(4,113 Views)
Thanks a lot.
It's ok.
(4,095 Views)