I am attempting to use the standard library single point read to measure current. Everything is working fine except that it is way too slow for our application. The 34401 datasheet says it can read 1000 readings / second and store 512 readings in memory.
we are getting about 3 readings / sec. I tried the Read (Multiple Points) to see if it was faster and I am still getting about 3 reads per second.
What can be done to speed this up? I would like at least 20-30 reads per second
Thanks
08-04-2015 12:29 PM
08-04-2015 12:33 PM
I'm not sure how to differentiate what you mean by the actual measurement. The current is quite low <100uA.
08-04-2015 12:36 PM
08-07-2015 01:53 PM - edited 08-07-2015 01:56 PM
@matt.spencer wrote:
I am attempting to use the standard library single point read to measure current. Everything is working fine except that it is way too slow for our application. The 34401 datasheet says it can read 1000 readings / second and store 512 readings in memory.
we are getting about 3 readings / sec. I tried the Read (Multiple Points) to see if it was faster and I am still getting about 3 reads per second.
What can be done to speed this up? I would like at least 20-30 reads per second
Thanks
With Auto Zero and NPLC both at the default settings you will get exactly 3 Reading per second if your primary power is 60Hz
10 power line cycles to integrate a "Zero Reading" followed by 10 PLC to integrate the current reading.
Options: