Realtime Waveform Acquisition LeCroy wavesurfer 44xs
ok, thank you all for the help so far.
I took the  "Edge Triggered Acquisition" example and added the "Configure Sample Mode" and put it on Roll mode. When running the application i get the error message: "<ERR> Selected sample mode is not available. Possible causes are:  incompatible model, sample rate or other horizontal system configuration."
the model should be compatible, right? so it must be caused by sample rate or other horizontal system configuration. Where can I see which settings are working and where can I adjust them?
Message 11 of 15
(2,087 Views)
Hi aelray,
Roll mode can be configured only when the scope is setup for long acquisitions.  Go to Timebase > Horizontal after you get the error message.  Is the "Roll" button greyed out?
You should call "Configure Timebase" before calling Configure Sample Mode (which I should have mentioned before!)
Best Regards,
Alan
Alan Blankman, Software Engineer
Teledyne LeCroy
800-425-2000
http://www.teledynelecroy.com
alan.blankman@teledyne.com
Message 12 of 15
(2,076 Views)
hi,
the "Roll" button was greyed out. With a bigger timebase it gets ungreyed.
So my hopefully last question is, that with the timebase configuration I  can set the time the measurement should take, right? And also I should adjust the "timeout" value accordingly, right?
thanks once again
aelray
Message 13 of 15
(2,038 Views)
ok, now i understand what timebase means, it's not for configuring the time of measurement ;). But now, after choosing a bigger timebase i got the problem that i get a waveform as result even if the measured waveform did not reach the trigger value...
Message 14 of 15
(2,027 Views)
Hi aelray,
Yes, Configure Timebase sets the scopes time per division and memory depth for the acquisition.  I don't understand the problem you describe below, however.  Can you email me your program, along with a screenshot?
Alan
Teledyne LeCroy
800-425-2000
Message 15 of 15
(2,018 Views)