Setting trigger level on AUX channel of DPO4032 with tkdpo4k driver
It looks for me that there is a bug in the trigger level limit check functions for the AUX INput of the tkdpo4k driver  for Labwindows/CVI (Version 4.3.1 , released  16-JUL-2009) when using a DPO4032 scope.
The instruments manual tells us (see  http://www2.tek.com/cmsreplive/marep/14938/071212103web_2009.02.10.14.36.48_14938_EN.pdf , page 77) : AUX IN: Trigger Level Range is adjustable from +8V to -8V.
However if I try to set the level to 1.5V  with tkdpo4k_ConfigureEdgeTriggerSource () it returns always IVI_ERROR_INVALID_VALUE.
 Looking into the source shows the following code in tkdpo4kAttrTriggerLevel_CheckCallback() :
   if (!strcmp (tSource, TKDPO4K_VAL_EXTERNAL))
        ext = 1.0;
    if (ext != 0.0)
    /*- Trigger source is External -*/
        checkErr( tkdpo4k_CheckRangeWithPrecision (vi, attributeId, value, -1.0 * ext, 1.0 * ext));
which limits the trigger level setting to +/- 1.0 V.
The same code  can also be found in tkdpo4kAttrTriggerLevelByCh_WriteCallback() and in tkdpo4kAttrDelayTriggerLevel_CheckCallback()  it is even limited to +/- 0.8 V (factor used in  tkdpo4k_CheckRangeWithPrecision ()  is 0.8 instead of 1.0) .
As minimal workaround I just changed the value of ext to 10.0 and the factor in  tkdpo4k_CheckRangeWithPrecision () to 0.8 in all variants and now it accepts the 1.5V without problems on the DPO4032.
Can someone @NI make sure that this problem is fixed in the next release of the that driver ?
However I'am not sure that my workaround is right for all scopes supported by this driver. I did't check all manuals. You might need to introduce a model dependant limit check here.
(4,557 Views)
Hi markus,
I am investigating the issue you request. The process may  takes one or two days and I'll tell you the result once we have a result.
Thanks.
(4,535 Views)
Hi Markus,
I have updated the source code and attached bellow. You can replace the file to the old one and test to see whether there are other issues.
Thanks.
(4,503 Views)
Accepted by markus kossmann
The problem seems to be solved now woth the updated version.
(4,477 Views)