05-30-2018 11:14 PM
Solved! Go to Solution.
05-31-2018 12:40 PM
Hi Lorenzo,
Since NI does not support the Keithley 2611A, I would contact Tektronix support for questions about this device.
Tektronix Support
https://www.tek.com/product-support
Best,
Misbah
05-31-2018 01:32 PM
Yes, an SMU is a source and measure unit. You can source either current or voltage while simultaneously sensing voltage or current. If you are going to automate your measurement then there's a LabVIEW driver for the Keithley 2611a - https://www.tek.com/source-measure-units/2635-software/keithley-series-2600-2600a-2600b-native-labvi...
Once you've installed the driver have a look at these two examples..
You will also need to enable remote sensing (4 wire mode) using the "\Configure\Measure\Measure Sense Mode.vi"
Craig
05-31-2018 02:07 PM
The 2600BS-901-01_C_Aug_2016_2.pdf manual has an example of your setup on page 2-57. You can use the SCPI commands to control the device by placing the SMU into 2400 emulation mode.