TDS 7054 query max and min measurement
I seem to be having difficulty querying for the maximum and minimum measurements on my TDS 7054 o'scope. I have both meaurements on and if I query MEASU:MEAS1:TYP? it returns MAX and returns MIN for MEAS2. yet when i try to query the actual value MEASU:MEAS1:MAX? it returns 99.000+36 which of course cannot be true. The strangest part is that the code works for the DPO4104. Any Idea what I'm doing wrong? I've attached 2 sets of code, 1 is how I turn the measurements on on the scope, the other is trying to query for max and min.
(4,530 Views)
Accepted by topic author JakeKing
You should be using ":MEASU:MEAS%d:DAT?" to get your data (put the number in for the %d).  You are actually requesting the maximium max value the scope has recorded.  It can keep some statistics on measurements.  Here are the commands I use to get the data (again, replace %d with the measurement number):
:MEASU:MEAS%d:DAT?
:MEASU:MEAS%d:UNI?
 The first command requests the actual data of the measurement and the second requests the units of the measurement.
(4,525 Views)
thank you for your quick reply
DAT is not an acceptable command on these scopes, but if i use VALue instead it works like you said it should.
Thanks again.
(4,519 Views)
"MIN" is not acceptable for your scope, you must use "MINI"
(4,330 Views)
MIN works fine thank you though
(4,322 Views)