TNT4882 embedded application DMA problem
Hi
  1.                      In DMA I don;t work with interrupts,then I write before beginning the DMA transfer 0 in IMR1,IMR2,IMR3 and 0x80 IMR0
  2.                      The value of register when the DMA stops:
        • ISR0                  0x05
        • ISR1                  0x10
        • ISR2
Message 11 of 13
(1,158 Views)
Hi
        • ISR0                  0x05
        • ISR1                  0x10
        • ISR2
Message 12 of 13
(1,158 Views)
I have some comments about the firmware:
At line 46 you write the RESET_CHIP command to the GPIB_AUXMR. This clears the GPIB_HIER and GPIB_MISC regsiters you configured in lines 43 and 44.
At line 72 you are setting the CPT_EN bit. This bit is used to detect undefined GPIB commands by asserting an interrupt. Do you need this functionality?
At line 66, you are setting REOS and at line 75 you are writing 0x0A to the EOSR. This means that whenever the TNT4882 receives the byte 0x0A it will stop. Is this the intended behavior?
My C skills aren't the best, but you have two different functions, outportb and OUTPORTB. They only seem to differ by the capitalization and I can't find the implementation of these functions. Do they differ in some manner?
I also have some comments about the register values after the failure:
ISR1 indicates that END is set. This may be due to the REOX setting I mentioned earlier.
I assume you are using the TNT as a non GPIB controller. Is that correct? Are you connecting to an NI GPIB controller (like a PCI-GPIB) and using the NI-488.2 driver for that GPIB controller?
I see you have an interrupt handler set up. What interrupts do you receive during the failing and passing cases?
Also, I don't quite understand the error condition reported during the failure. Is it just a timeout?
Some of the other regulars on this board are much more knowledgeable about TNT4882 firmware, so maybe they will have some more ideas.
Message 13 of 13
(1,156 Views)