USB-232/2 shows COM ports but no VISA resources found!
How is the agilent instrument connected? You would only need the usb driver if it's connected like that.
At least set NI-VISA as primary.
Message 11 of 19
(4,051 Views)
You can have both NI and Agilent VISA installed but only one can be active at a time. NI VISA must be active in order to communicate with Agilent Device with NI
LabVIEW or MAX.
You have to follow the sequence described in the below link to do that and it works.
http://digital.ni.com/public.nsf/allkb/3B3626D9C1F999218625694200791AD7
Message 12 of 19
(4,041 Views)
The Agilent instrument is connected via GPIB. The Agilent communication has no problem. I followed the steps here http://digital.ni.com/public.nsf/allkb/CC76C6A93C0F3623862574E50047E269, made sure that NI-VISA is the primary VISA. I do have the NI VISA and NI Serial to USB visa installed.
Message 13 of 19
(4,021 Views)
Hi, I have followed the steps in the link. The Agilent communication has no problem, but I still can not find the temperature controlelr as a VISA resource.
Message 14 of 19
(4,018 Views)
Are you using an agilent GPIB controller?
Message 15 of 19
(4,017 Views)
The Agilent impedance analyzer is connected through a USB/GPIB interface. The control in computer is only by LabView.
Message 16 of 19
(4,014 Views)
A simple yes would have been enough.
Message 17 of 19
(4,004 Views)
Accepted by WenyanJiang
I strongly suspect the installation issue.
If you do not get any other solution, I would suggest uninstalling everything and installing the software’s in following order
1. Install LabVIEW
2. NI Device Drivers
3. Install Agilent IO Library and make sure you do not set Agilent VISA as an Primary during the Installation
4. Configure NI Max to communicate with the third party IEEE 488 by enabling NI VisaTulip.dll
Message 18 of 19
(3,992 Views)
Hi! I got a new hard drive, and installed new operation system, labview and other drivers on it. Labview is now communicating with temperature controller! Thanks for the help.
Message 19 of 19
(3,967 Views)