05-18-2012 08:10 PM
my connection configure are little complicated: I use laptop Latitude E6510 (which has no serial port) set in docking station. I use USB cable connection docking station and NI USB to RS-232 (2 port) device. then use 9-pin cable connect between converter and Keithley 2702. I can operate instrument in multi-switch scan without any problem. But when I use MAX to search the instrument which always get error -9996.
My question is: Does the USB to serial convert device have to connect PC directly?
Solved! Go to Solution.
05-18-2012 09:08 PM
05-18-2012 09:10 PM
05-19-2012 12:12 AM - edited 05-19-2012 12:13 AM
MAX="Measurement & Automation Explorer"
I have to use the convert because my laptop does not have serial port
05-19-2012 02:31 AM
Hi
There is no problem with the serial port or converter when it works. A usb connection in the docking station is a connection in the machine itself for the software. I have also an e6510 and see no problem.
MAX will not find the k2700 becasue MAX does not scan the serial ports.
05-22-2012 09:41 AM
Hi I also have a similar problem
I have an USB to R232 converters and im trying to communicate with a Varian AGC-100 with attaced Manual
Im using Measurement and Aquisation explorer from NI
The problem i have is that i just get some random pressures no matter which command i send to the device
using Query or write and read
has anyone an idea what the problem could be
with kind regards Venu