03-28-2006 01:54 PM
03-28-2006 01:55 PM
03-29-2006 07:41 AM
03-30-2006 12:10 PM
03-30-2006 12:43 PM
03-31-2006 03:15 PM
Dennis,
Q: Is it the Keithley or scope that is supposed to be in differential mode? A: The CSA8000 is what is in differential mode. When the program is activated the scope sets up and functions as it should except that the Keithley switch does not operate as a 1x48 multiplexer, it operates as a 1x96 multiplexer so the scope does not truely operate in differential mode. Q: Can you post the section of code that does the mode change? A: That would be a huge file however, prehaps I can give you a better explanation. The setups, wether differential or single-ended, are separate setups that the user selects. Based on that mode selection, the program recalls the respective scope setup and the first memory location of the Keithley switch is called through an initialization file - single-ended is location 1 and differential is location 100. In the differential case the scope sets up in differential mode, but the Keithley switch does not operate as a 1x48 multiplexer. It will switch 48 times in a 1X96 configuration. That leads me to believe that the problem may be related to the GPIB - but I've been wrong before.
The big challenge for me is that the tester is now off-site and I don't readily have available access to it.