Hello
I am trying to connect RS232 (9 pin) via serial ports..I also have Measurement & Automation Explorer (ver 5.1) to monitor and configure Serial/parallel connections.
The problem is that when I query, it looks like my device is not properly recognized. On "Open VISA Test Panel", I tried to enter commands on "Input/Ouput", I only get time-out errors. Additionally, on the same panel, I looked up "View Attributes", and everything looks OK except the very last row where it says "Error VISA: (Hex 0xBFFF001D) The specified attribute is not defined or supported by the referenced resource".
How can I ensure that my device is properly recognized? Is my COM1 port activated at all?
Thank you
12-01-2012 12:20 PM
12-01-2012 12:58 PM
Reading your messages, I have a few more remarks and questions:
1. I use a High Resolution Motorized Microsmanipulator (WPI-HS6-3M). The manipulator is driven by a stepper motor controller (Marzhauzer, MCL). The only interface on MCL is RS232.
2. COM1 setting: I looked up the MCL manual, and the only thing I was able to find is baud rate, which is set to 9600 (this is only one-page manual). I have no idea what kind handshaking it uses nor flow control settings I need to use. How do I find out this?
3. RS232 cable: 5 is common and grounded with 2&3 and 7&8 handshaking. Or at least that's what I have found by checking it with a multimeter.
4. How do I check if my device is appearing in MAX?
5. One thing I found was that even after I disconnect MCL, I still see COM1 showing up on the pannel. In other words, as far as the COM1 display goes, I get the same things whether or not my device is connected. This is quite different for the other USB-Serial device, which appears to be at COM2 port. There, if I connect, the display on the pannel is accordingly updated with the correct device name.
Thanks for your help
12-01-2012 01:19 PM
12-03-2012 02:37 PM
I have confirmed with the vendor on all the relavant parameters (Baud rate 9600, no parity, 8 data bits, and 1 stop bit). The device also uses the standward RS232 cable.
I am still getting an error on one of the attributes (see attached). Can you give me some advice as to how to address the error in red?
Thank you
12-03-2012 03:24 PM
As I said, that is irrelevant. The property is only for NI serial devices.
12-03-2012 03:44 PM
Can you make your advice/suggestion more constructive (and specific)? What are some possible things I can try?
12-03-2012 04:12 PM
For that property/attribute, you do nothing at all. Irrelevant means you ignore it.
If you have timeout errors, you have to do what I said. Check the com settings, check the cable, check the commands. You have not provided your code, the manual, the type of cable you are using (no such thing as a 'standard RS-232 cable), the commands you are sending, the termination character you are using, etc.
12-03-2012 04:39 PM
1. com settings: As I said, it's Baud rate 9600, 8 data bits, 1 stop bit and no parity. See attached.
2. RS 232 cable: I have DB9(male/female) with null modem (cross-wired) configuration. I also tested its connectivity (2/3 and 7/8 and 5 grounded)
3. commands/codes: I simply query the ideniticaiton of my current device. See attached.
4. Manual: There is no manual for the current device. The only thing I found was a one-page summary spec. Here it is: http://www.marzhauser.com/en/products/controllers/controllers-for-micromanipulators/mcl.html
It's the last product MCL 3 with display.
Let me know if there is anything specific I can try given the parameters/conditions I have provided you with.
12-03-2012 05:36 PM
You just can't make up commands to send to the device. You HAVE to get the programming manual from the vendor. The *IDN? command is an IEEE-488.2 standard and it is not at all common for serial instruments.