VISA address of instrument cannot be refreshed

Solved!
Go to solution

I used labview 8.6 in Winxp to control instruments such as keithley 2400, 2602, SR830 to perform measurement. Recently, this computer affected by virus, so we reinstalled the system, changed to Win7 system. We can find the instrument in MAX, and also can communicate with instrument. But we failed to control the instrument in Labview 8.6. It seemed that the insrument cannot be found in Labview. After we installed the Ni-VISA 5.1, the problem is still there. Can someone tell me how to do?

 

    Thanks

 

 

0 Kudos
Message 1 of 3
(3,076 Views)
Solution
Accepted by topic author huafeng

Hi,

 

http://digital.ni.com/public.nsf/allkb/B972242574D4BB99862575A7007520CB

 

LabVIEW 8.6 is not listed as being compatible with window 7.

 

You probably need to upgrade to a LabVIEW which is compatible

Regards
Ray Farmer
0 Kudos
Message 2 of 3
(3,041 Views)

Thanks!

 

 I reinstall the WinXP system, NI 488.2, then everything is OK.

 

 

Message 3 of 3