What is the purpose of a period in ibcmd calls?
I'm converting an old GPIB calibration script and am seeing periods in the oddest of places and cannot determine the reason. Can someone explain?
For ibcmd I can understand this...
ibcmd(GPIB0, "?U!", 3 (0x3))
ibwrt(GPIB0, "MEASU:MEAS1:V...", 17 (0x11))
ibcmd(GPIB0, "_?", 2 (0x2))
But what about this one?
ibcmd(GPIB0, "?_5.", 4 (0x4))
ibcmd(GPIB0, "A", 1 (0x1))
ibrd(GPIB0, "P", 1 (0x1))
ibcmd(GPIB0, "._?", 3 (0x3))
It appears to be only seen when working with the board.. the Multiline interface message ends with a period then the untalk unlisten calls begin with one.
What is the reason behind this additional period? I would think that it would overwrite MLA/MTA's in the same command.
(3,791 Views)
For the first ibcmd("?_5."): The result will be that devices at GPIB addresses 21 and 14 will be addressed to listen. Keep in mind that on GPIB there can be multiple listeners, but only a single talker.
For the second ibcmd("A"): The result will be that the device at GPIB address 1 will be addressed to talk.
The last ibcmd("._?") does not make much sense. The UNT(_) and UNL(?) will override the previous MLA14(.).
Hope this sheds some light on your problem.
Craig A.
National Instruments Engineer
(3,791 Views)
Ding! The lightbulb just went off. Thanks Craig.. now what I'm looking at makes perfect sense.
(3,790 Views)
Two additional questions.
>>For the second ibcmd("A"): The result will be that the device at GPIB address 1 will be addressed to talk<<
Where do you get 1 from the A?
Is the P in the ibrd an indication of the set bit masks?
(3,790 Views)
An ASCII A, or HEX 41 corresponds to MTA1. Look in NI-488.2 Help for "IEEE 488 command messages" for a full table of command message definitions.
Is this the result of looking at an NI-Spy capture, or is this how the actual code was written? If this was from the Spy capture, the "P" was the data read from a device.
Craig A.
(3,790 Views)