What's the GPIB for spectrum analyzer amplitude attenuation?
Hello
I am writing a test automation, and I need to do the following by GPIB programming:
Press AMPLITUDE Y Scale, Attenuation, 0, dB.
The device is Agilent E4407B spectum analyzer.
I searched the internet and researched and cannot find the GPIB programming string that commands amplitude attenuation.
(I did find all the other commands I need for this automation.)
Actually, the test plan that I am implementing in programming, says that this one is only available via keyboard entry.
Could it be? All the keyboard commands have corresponding GPIB commands besides this one? Why?
Perhaps it is because the same effect can be achieved by a different GPIB command? What is it then?
Thanks!
Tova
Message 1 of 12
(5,064 Views)
I seriously doubt Agilent left out a commmand.  You just have to figure out where it is.  Have you tried downloading the LabVIEW instrument drivers from NI?  You can often find little gems in those.
Message 2 of 12
(5,058 Views)
No, I didn't download the labview instrument drivers from NI.
I do have NI drivers and NI GPIB explorer.
Is there a list of all existing GPIB string commands there is labview drivers? Or how could these drivers help me?
Tova
Message 3 of 12
(5,056 Views)
I guess I should ask what language you are using in order to do this testing.
Message 4 of 12
(5,054 Views)
I think this is the command you are looking for:
:POW:ATT ##;
Message 5 of 12
(5,053 Views)
I use TCL programming language.
But before using a new GPIB command in programming I try it out first using NI GPIB explorer which does not invlove any programming or programming language.
Message 6 of 12
(5,046 Views)
Doesn't work  😞
SA displayed "Undefined header"
I tried it with NI GPIB explorer. (no programming)
Message 7 of 12
(5,045 Views)
Add the correct SCPI header. I think it's UNIT:
In any case, you should be able to quickly look up this command in the manual. And, if you continue to have problems with the command, you should contact Agilent. This board is for NI hardware/software and you are not having problems with either. Agilent makes the instrument and they can help you better than anyone else.
Message 8 of 12
(5,036 Views)
Hi
I contacted agilent and they gave the solution:  SENS:POW:ATT <val>
Thanks  🙂
Tova
Message 9 of 12
(5,004 Views)
Now I realize that in your reply above an emoticon "ate" ':' and 'P'  !
As in my reply.
You were right then. But I tried without the 'P'.
Message 10 of 12