Wrong reading operation with viRead() - CVI
I am trying to read trace points through a spectrum analyzer (Agilent (HP) 8563E). It is pretty old, by the way.
When using the test panel to get the traces, I get all data points from just one click on the "Read" button.
However, using C on CVI, I am getting only 11 bytes per reading operation:
sprintf(buffer_file,"TDF P;TRA?\n");
status = viWrite(instr, (ViBuf)buffer_file, strlen(buffer_file), &retCount);
/* for (bit = 0; bit < 50; bit++)
status = viRead(instr, (ViBuf)buffer_file, strlen(buffer_file), &retCount);
printf("%s\n", buffer_file);
I tried to put a for loop just to see what happens. When performing successive read operations, I get 11 bytes per reading, while I was expecting to get all traces points by one reading operation. The print screen of the results are attached.
Also, I was clearing the buffers before writing/reading.
I am using GPIB (VISA). Any help will be appreciated.
Thanks!
(5,006 Views)
It sounds like you have to disable the term character on read property. There are 2 CVI drivers for that instrument. Have you tried either or looked at the source code?
http://sine.ni.com/apps/utf8/niid_web_display.model_page?p_model_id=1916
(4,990 Views)
Hi Dennis,
I tried disabling it, but had no success. I'm actually pretty curious about why it's not working.
I got it working with the drivers.
Thanks for answering!
(4,971 Views)
I am having the same problem with another instrument (ANDO AQ6317B Spectrum Analyzer). When reading the trace, I just get 11 bytes instead of all trace points.
Is there any parameter that can limit the number of bytes that are being read? I already disabled the termination character and nothing changes. The termination character used by the instrument is \r\n.
Felipe
(4,938 Views)
Since the driver for the 8563 works, I can only suggest you look at its source code to see how it's configured. The same setup should work for your new instrument.
(4,932 Views)