Yokogawa UP550 RS-485 Communications
Wanted to get this info available for a search: Yokogawa UP550 RS-485 Communications
A terrific earlier post for the UP351 was available and it includes very helpful info for the UP550 model,
and a link for the Yokogawa Document  "GREEN Series Communication Functions"  IM 05G01B02-01E
http://forums.ni.com/t5/Instrument-Control-GPIB-Serial/Communication-Protocol-with-RS485-and-Yokogaw...
(6,206 Views)
slbLV,
I am not sure what you mean by having it "available for a search."
The link you posted does not lead to a valid page.
Michael L.
Sales Engineer
(6,156 Views)
Probably this: http://forums.ni.com/t5/Instrument-Control-GPIB-Serial/Communication-Protocol-with-RS485-and-Yokogaw...
Since the old thread was obviously searchable, I don't understand the post.
(6,151 Views)
Dennis_Knutson,
Thank you for the correction in the Link.
Last week I was searching for UP550 and the only item that displayed was the referenced Link,
and only because some else asked a question in the UP351 thread concerning the UP550.
My attempt with this new thread was to focus on the UP550 and show that it was very useful information.
Was that action (creating a new thread) NOT a good idea ??
(6,145 Views)
Since the old link did reference the UP550, I just didn't see the point but it's really no big deal.
(6,140 Views)
More Details:
Note there is NO CheckSum due to the use of the Communication Mode    PSL1 = 0
The CORRECT Example for the INF Command is:
\0201010INF6\03\r
[hex: 0230 3130 3130 494E 4636 030D ]
which provides this feedback (may vary on some of the Data Specifics):
0101OKUP550-01V02.R26 0001002502080000
[hex: 0230 3130 314F 4B55 5035 3530 2D30 3156 3032 2E52 3236 2030 3030 3130 3032 3530 3230 3830 3030 3003 0D ]
See this link for more details on the Functions and Programming :
 http://www.yokogawa.com/us/products/controllers-pid-plc-rtu-pac-hmi/controllers-indicators/program-c...
Instruction Manuals
(6,025 Views)