communication with tektronix 3014B via TCPIP
I am trying to get the waveforms from scope tektronix 3014B via enthernet connection (TCPIP protocol). I tried the the instrument drivers, but it seems they are for RS232 or GPIB only. Then I tried to make a simple program to communicating with the scope in the same local network.
However, it doesn't work and report overtime. Anyone has clue what is wrong with this vi? Thanks!
Message 1 of 14
(7,428 Views)
The download page says they are for Ethernet as well as serial and GPIB. Have you tried to add the instrument in MAX as a VISA TCP/IP resource? Can you ping the instrument?
Message 2 of 14
(7,418 Views)
Hi Thanks.
Now I've added the port address in MAX and configured the alias name. After that I can see the alias name appeared in the draw down list of the resource name of the example. However, I still cannot run it. Still, every time, it reports time out.
I am using the driver in this page
http://sine.ni.com/apps/utf8/niid_web_display.model_page?p_model_id=163
And I've tried both the plug and play and ivi style drivers. By the way, I am using Labview version 7.1.
Thanks.
Message 3 of 14
(7,410 Views)
Hey bo,
Now that you are using the instrument drivers we provide, can you post a screenshot of the VI that you're trying to run, or the VI itself? When the timeout occurs, what function is causing the timeout error? If the Tektronix device is configured as a VISA TCP/IP Resource, you should be able to open up the VISA Test Panel for the device. Try opening up the test panel and communicating with your instrument there. Thanks!
Justin E
National Instruments R&D
Message 4 of 14
(7,365 Views)
If you are using external trigger - you must have a probe on 1 of the channels this usually causes of time-out error
Message 5 of 14
(7,354 Views)
Thank all for your input.
I tried to use function generator to input a sine wave on ch1 and the synchronization is on ch4. I tried to trigger on ch1 and it doesn't work right either.
I am attaching screen shot of the max configuration and running window of the test program with error message and the ethernet interface of the scope showing the output signal. Also, the vis are attached and the program I run is tktds3k Single Channel Waveform Acquire Example.vi
Thanks!
Message 6 of 14
(7,339 Views)
I would recommend running the example in the highlighted mode to check where exactly it fails.
Also, If you set the "trigger level" too high or too low (causing the acquisition not to trigger), the timeout error will occur. In this case, the oscilloscope displays the message "Trig?" on the instrument display, which signifies that the signal did not trigger the acquisition. You should change the trigger level. If this does not help, you should also try to use the "Initiate Acquisition.vi" and then the "Fetch Waveform.vi" instead of the "Read Waveform.vi" that is used in the example.
Have you tried to run any other example or to use a LabVIEW project-style driver (supported only in LabVIEW 8.0 and later)?
Message 7 of 14
(7,318 Views)
Thanks for your advices.
Unfortunately it did not help either
I found the error happens in the initialization.
I tested *IDN?\n in MAX, Here is the message the MAX displayed:
HTTP/1.1 505 HTTP version not supported. Expected HTTP/1.1, received ...Date: FRI MAR 13 09:10:02 2009..Pragma: no-cache..MIME-Version: 1.0..Connection: close..Content-Type: text/html....<HTML><HEAD><TITLE>HTTP Version Not Supported</TITLE></HEAD><BODY BGCOLOR="#ffffff">..<H2>505 - HTTP Version Not Supported</H2><P>HTTP version not supported. Expected HTTP/1.1, received ...</BODY></HTML>..
Looks like the scope supports version 1.0 but the labview only supports version 1.1
Message 8 of 14
(7,301 Views)
The problem is solved. What I did is downloading the latest version of TekVISA from tektronix.com and installed it in the computer running labview.
In the installation process, there is a step asking whether to substitude ni visa to tekvisa. I selected yes. I don't know the impact of that to other labview visa applications yet.
After installation, the example runs fine.
Bo
Message 9 of 14
(7,294 Views)
Hey Bo,
Glad to hear that you got the issue fixed!
Both NI-VISA and Tek-VISA have been tested to work with LabVIEW, so hopefully you won't have too many issues with making this switch.
Justin E
Message 10 of 14
(7,284 Views)