Dear all,
For my project, I have to measure a difference of pressure of gas. As the range will be quite important, we will use 4 transducer, and change with a physical valve ( 1 transducer is enough for 1 test).
The material I have : - transducer (10V supply, give a signal back of 0-100mV depending on the pressure)
- data acquisition card NI PCI-6023E
- Block connectors CB 68 LPR mounted in in-house made box.
This box also contain the power supply - a transfo from 220V AC to 10V DC. The box contain also the female connection for the transducer, with old 5 pin DIN connector (http://en.wikipedia.org/wiki/DIN_connector) and the wires to the CB68LPR. To each connector, I have the possibility to put a multimeter in parallel.
Wiring connection : all female DIN power supply are in parallel, the voltage stay roughly good when connecting 2 transducer. Each AI and corresponding ground wires are twisted together until the CB 68 LPR, and all ground goes to separate AI GND. Previous version of the box, I had 100mV ground put together to the port 64 (AI GND), but I had the same problem.
I already wrote a VI which capture the data without too much problem ( I have to do an average as the signal seems to have 50Hz noise- RSE mode) as far as I connect 1 transducer to the box, no matter which analogue input canal. Also, having 2 transducer, but the cable NOT connected to the computer works fine.
The problem appears as soon as I connect 2 transducer AND the cable to the computer. In this case, I have stupid signal, often more than 100mV on the multimeter, I do not dare to let it enough time to check what LabView see.
If somebody has any idea what happen, and how to solve it, I would really appreciate.
If any more information could help understanding the problem, please ask! It can be something easy, as I am far away of my background.
Already thank you for reading all this.
Regards
Renisaac
05-20-2010 09:26 AM
Dear all,
I did not progress at all 😞
I checked that the computer is correctly grounded, and I changed the acuisition card (we had a PCI-6023E spare), without any changes.
I do not know what else to check !
Does anybody have an idea, even the most crazy one? Any indication of a direction to investigate will be really appreciated!
Cheers
Renisaac
06-01-2010 07:44 AM
Hi Renisaac,
exactly the problem I cannot define.
What would help to define the problem, is an image overview of the applciation, and the way all signals are connected to the different elements in the measurement application.
Regarding grounding you can read the following document. It provides valuable information on the grounding subject:
If you need additional help, please try to provide a scheme of the application and connections. That will clarify quite a lot.
Good luck!
Regards,
Peter S