issue talking with 2 devices connected via GPIB
Hmmmm
Sounds like a really old 488 design circa mid 60s-or 70s  The product brochures state a 488 link- (not the current 488.2) so I suspect a "non standard" implementation.
Post a screenshot of the MAX GPIB devices- this will give us a clue how MAX sees the instruments
Post a drawing of the bus - so we are all on the same page
If you can, post a link to the equipment operating manuals- I can't seem to find them online
Message 11 of 16
(1,445 Views)
Is the 488.2 not backwards compatible?
So the equipment is definitely old. I found a link to one manual and site to get the spec sheet for the other.
Model 1287: http://mm.ece.ubc.ca/mediawiki/images/3/35/1287_manual.pdf
Model 1252A: http://www.solartronanalytical.com/Pages/1252AFRAPage.htm
Attached a screenshot of the MAX after scan for instruments.
Also the GPIB cable has a tag on it that reads:
763507-02 Rev1
Type-X2
Length 2.1meters
2002 10 28
Let me know if I got everything you asked
Message 12 of 16
(1,434 Views)
Hard as it is for me to believe, each of those instruments has two separate GPIB addresses according to the manuals. That accounts for for MAX seeing four. They are all, obviously powered on and I cannot account for the problems you are seeing. A call to the vendor might be in order.
Message 13 of 16
(1,415 Views)
WOW-
OK the dual addressing wasn't too uncommon back in the day.  Especially when the unit was designed to dump to a HP plotter.  I'm not sure you are going to get there using the VISA calls since VISA assumes a "IEEE-488" COMPLIANT interface and this is what I would call a "GPIB style" interface (probably on a interface card developed in-house expressly for the Solaratron equipment.)  Again, this was common back in the day.
Dennis offers good advice!  Time to call the vendor, explain you would like to use their obsolete interface with modern instrument control software, and work closely with their application engineer.
I've been-there-done-that a few times myself.  Its painful but, the equipment manufacturers I've worked with have all been "customer focused" and in every case the manufacturer spun a revision of the interface to bring it into IEEE-488 compliance.
Message 14 of 16
(1,393 Views)
Hi
In this case I would try each of the four addresses and use the two working addresses.
I have used the solartron with visa for sure and although you have to watch out for timeouts when measuring slowly it should work.
The other instrument I did not use.
greetings from the Netherlands
Message 15 of 16
(1,380 Views)
@Albert Geven wrote:
Hi
Of course I could be mistaken-  Thanks Albert  The even addresses are the right ones-
unless the 1's address bit is on then the device won't work right.
Message 16 of 16
(1,375 Views)