I'm trying to control several mass flow controllers that are attached to a Alicat BB9 serial hub. The hub is then connected to the computer via RS-232 serial. The computer this is being done on does not have a serial port, so a serial-to-usb adaptor is used.
When plugged in, the computer (running XP) recognizes the serial-to-usb
and asks for a driver. I have installed a driver from the manufacturer
of the adaptor, downloaded from the internet. The same driver is used
on another computer (running Vista) with no problems.
I am able to verify that the device is installed and working
correctly through Device Manager. Also I am able to view which COM port
the device is on. Opening NI's Measurement and Automation Explorer
(MAX) I am able to see the COM port listed under serial devices. I am able to validate port settings and open a VISA session with the device. However, any time I try to read from the device it times out. I get the error xBFFF0015. I have been able to up the time out setting to 5000 and 10000 with no change.
When running a VI that is used around the lab to control the same setup, I can trace the error back to a VISA read subVI. The 'read
buffer' returned doesn't have a value and triggers an error. I can plug
the USB into a laptop that has the same VI and it will run with no
trouble. The 'VISA read' subVI does return a value when running on the
laptop.
I am running LabView 8.5 and VISA 4.4.
If anyone has a suggestion that could get this working I would appreciate it greatly. Also, I'm a grad. student working on this and my understanding of LabView isn't terribly great, so dumbed-down responses would be nice 🙂
Thank you for your help.
Connor
Solved! Go to Solution.
03-05-2009 02:48 PM
Hey comoyer,
0xBFFF0015 is generally a serial timeout error, so upping the timeout setting is a good first step to try and resolve your issue. Have you tried communicating with your device through Hyperterminal or through a test panel in MAX? To get to the VISA Test Panel, open up MAX and navigate to your COM port. You should be able to click on "Open VISA Test Panel" to open a test panel for this COM port. Try writing and reading to/from your device with these applications to see if you still get the same error.
When you plug the serial-to-usb adaptor to your laptop, is everything configured the same as it is on your desktop? What operating system is on your laptop?
03-05-2009 04:12 PM
03-05-2009 04:15 PM
07-04-2009 07:29 PM
Hi, Connor,
I'm using the same setups with yours and I have difficulty in figuring out communication problem between Alicat mass flow controllers and computer.
Can you provide your email that I can contact to? I'd like to ask some questions. My email is swsyoon@gmail.com.
Thank you.
Yoon