05-08-2013 02:00 AM
Hi
I want to do test with Agilent 8510A network analyzer. I was replaced its sweep oscillator 8350B by the Rohde & Schwarz SMP 22 signal generator. I changed SMP 22 GPIB address to 19. but when I want to set center frequency or other similar parameters in HP 8510, they don't set ( be unchanged) and a syntax error (eroor code -102) appears on Rohde & Schwarz SMP 22 display. I would really appreciate any help. Thank you Yasamin Gharib
05-08-2013 02:08 AM
Hi,
the SMP is controlled by 8510? What command is sent to SMP? Error -102 means: Syntax error. The command is invalid. Example: The command contains block data the instrument does not accept.
Milo
05-08-2013 02:18 AM
Hi
I tried to change center frequency and span frequency in HP 8510, but the values didn't change ( I couldn't set desire values for them) and error appears on SMP 22 display.
05-08-2013 02:48 AM
I'm affraid this will not work - 8350B has different SCPI commandset than SMP. If you cannot tweak the 8510 to send different commands there's not much else you can do 😞
05-08-2013 05:47 AM
I agree that 8350B has different SCPI command set than SMP. Is it concluded that the two instruments ( SMP 22 and 8510) are incompatible?
05-08-2013 07:48 AM
I tried to check the Agilent doc, but they have only scanned version so it's disaster to find something.
05-08-2013 04:13 PM
05-08-2013 04:57 PM
The driver won't help him much with the problem stated as the 8510 only uses a certain set of Agilent Signal Generators that are controled by the 8510 thru the System GPIB Bus, not the regular GPIB bus. So the sig gen he is using is not on the cpompatible list of devices usable for this instrument.