4-probe measurement using Agilent B2912A
I am trying to set B2912A Source meter to measure resistance in 4 probe method.
I tried looking into the device driver's tree.vi, however, couldn't find any way to configure the device for 4 probe mode and to gather data in that mode.
If I were to measure it manually. I would set it to 4 probe mode, then set the resistance range to V/I mode, set the source voltage (current) something like 2-5V depending on expected resistance and based on the sense current resistance value would be read out.
I'm very new to LabVIEW I have never interfaced any instrument so far, I have attached the vi that I have made so far, any help would be greatly appreciated.
(3,496 Views)
Accepted by topic author iamashwin99
What you are looking for is more commonly referred to as Remote Sensing.  There is a VI in the Keysight B29xx LabVIEW driver for setting that up or SCPI command is ":SENS:REM ON;"
Depending on what you are measuring its usually best to source current and measure voltage.  As such Keysight has an automatic resistance measurement mode where it sources increasing current until a voltage limit is reached (2V default).  Here's how I do that with remote sensing.
(3,456 Views)
Thanks for the tip on Remote Sensing block.
This worked for me
(3,442 Views)