4 wire RTD measurement
Any module can do 4 wire RTD measurement with thermal resistor around 10k to 20k ohm, total 20 channels, thanks.
(3,435 Views)
With such high channel count for RTD, you should use SCXI form factor.  See link below.
https://www.ni.com/en-us/support/model.scxi-1102.html
Kudos and Accepted as Solution are welcome!
(3,434 Views)
Can it take 20k ohm input, thanks.
(3,427 Views)
Below is the manual for the module that you need.  The maximum resistive load is 100kohm, os it will be able to take 10kohm.
https://www.ni.com/docs/en-US/bundle/373074c/resource/373074c.pdf
(3,415 Views)
May I know if the scxci module can accept 4 wire input module such as NTC thermistor (10K@25DegC).
Thanks.
(3,404 Views)
Did you read the manual which was linked in an earlier post?  Look at page 4-2 of that manual.  If you do not understand how the equipment works, perhaps you would be better off hiring a consultant to put together your system and to advise you on how it works and its limitations.
Lynn
(3,391 Views)
Accepted by topic author alex.
Hi Alex,
Yes, it would work.
Yik
(3,375 Views)