03-23-2011 11:08 PM
Solved! Go to Solution.
03-23-2011
11:18 PM
- last edited on
03-14-2025
05:23 PM
by
Content Cleaner
With such high channel count for RTD, you should use SCXI form factor. See link below.
https://www.ni.com/en-us/support/model.scxi-1102.html
------------------------------------------------------------------
Kudos and Accepted as Solution are welcome!
03-24-2011 12:04 AM
Can it take 20k ohm input, thanks.
03-24-2011
09:19 AM
- last edited on
03-14-2025
05:24 PM
by
Content Cleaner
Below is the manual for the module that you need. The maximum resistive load is 100kohm, os it will be able to take 10kohm.
https://www.ni.com/docs/en-US/bundle/373074c/resource/373074c.pdf
03-24-2011 08:03 PM
May I know if the scxci module can accept 4 wire input module such as NTC thermistor (10K@25DegC).
Thanks.
03-25-2011 07:27 AM
Did you read the manual which was linked in an earlier post? Look at page 4-2 of that manual. If you do not understand how the equipment works, perhaps you would be better off hiring a consultant to put together your system and to advise you on how it works and its limitations.
Lynn
03-25-2011 11:50 AM
Hi Alex,
Yes, it would work.
Yik