4072 capacitance verification
4072 capacitance verification: The 270pF verification fails. The calibration executive verification program does not open, short and null the lead resistance in the automated program. I put in the actual capacitance measurement with the test leads and capacitor and still get a failure at 270pF. When I test the 4072 manual on the front panel after performing the short, open compensation and null it reads correctly. Should a compensation and null be added to the automated procedure.
(2,649 Views)
The automated procedure in Calibration Executive should be implementing the steps shown in the calibration document for the 4072 exactly:
https://www.ni.com/docs/en-US/bundle/4070-4072-cal-pro/resource/4070-4072-cal-pro.pdf
It does not appear that those operations are performed in the procedure.  I believe the short,open,offset compensation are more of a correction than a calibration for the device.  While performing those operations will always give you more accurate readings, they do not adjust the general calibration constants on the device.
Jon F.
Technical Support Engineer
(2,600 Views)