65582 Error occurs in the middle of an acquisition on FPGA
I've got a 9234 acquiring data at 51.2kS/s on a 9074. The loop on the FPGA responsible for acquiring data from this module terminates unexpectedly after about 300 seconds with this error from the IONode: 65582
Possible reason(s):
CompactRIO:  (Hex 0x1002E) The operation did not complete because the acquisition has not started.  Use the Start and Stop channels to put the module in and out of acquisition mode.
Obviously, I used the Start node to start the acquisition, so what else could cause this error to occur? I check to see that the loop on the FPGA was keeping up with the sample rate, and that does not appear to be a problem.
Any other thoughts?
hmm, I just swapped the 9234 module with another identical module, and I'm at 600 seconds and counting. Could this be an indication of a bad module?
Chris
(3,657 Views)
(3,645 Views)
Yes. I am using NI-RIO 3.2.1, and I did start the acquisition. Like I said, it works fine for about 300 seconds before erroring out.
thanks,
Chris
(3,637 Views)
In order to check if you have a bad module I would recommend using the Module Specific example for the 9234 (do a search for the number in the example finder). This code is tested and proven so if it doesn't work then you may have a dammaged unit. At that point go to NI.com/support and set up an RMA so you can return or repair the device.
-Hunter
Message Edited by Hueter on 01-07-2010 05:27 PM
(3,623 Views)