8846A Fluke DMM has a proplem with the reading and it shows as bellow
Voltage DC Open Loop Value
Range  Unite Reading  Uint
100 mV OL mV
1 V 0.128 V
10 V 0.128 V
100 V 4.08 V
1000 V 3.95 V
Current DC Open Loop Value
100 µA OL µA
1 mA 0.41 mA
10 mA OL mA
100 mA 41.5 mA
400 mA 41.5 mA
1 A OL A
3 A OL A
10 A 4.22 A
Resistance Short Loop Value
10 8.08
100 40.73
1 KΩ 0.0406 KΩ
10 KΩ 0.0406 KΩ
100 KΩ 4.04 KΩ
1 MΩ 0.0039 MΩ
10 MΩ 0.039 MΩ
100 MΩ 0.039 MΩ
1 GΩ 0.039 GΩ
The Reaging of the applied  value on the range adds to half the applied value + the readings in the above tabels
For Example at range 10 ohm the value show up 8.08 if we applied 1 ohm it will equal (8.08 + (applied Value /2)) and it shows 8.58 ohm
For Example at range 100 ohm the value show up 40.73 if we applied 100 ohm it will equal (40.73 + (applied Value /2)) and it shows 90.73 ohm
For Example at range 100mA the value show up 41.5 if we applied 10 mA it will equal (41.5 + (applied Value /2)) and it shows 46.5 mA
Voltage AC Open Loop Value
100 mV 7.3 mV
1 V 0.0072 V
10 V 0.0072 V
100 V 0.0475 V
1000 V 0.638 V
Current AC Open Loop Value
100 µA 0.09 µA
1 mA 0.0006 mA
10 mA 0.005 mA
100 mA 0.06 mA
400 mA 0.575 mA
1 A 0.00087 A
3 A 0.0057 A
10 A 0.0062 A
(648 Views)
Since you posted in the LabVIEW forum, we must assume that your question has something to do with LabVIEW. Please explain.
So far, all you show are some big tables. Can you point out what you perceive as problem?
(645 Views)
As I mentioned earlier, the DMM (Digital Multimeter) seems to be showing incorrect readings. Even when no wires are connected to the mA or A jacks, the meter still displays a reading of 4.2 A on the 10A range.
(607 Views)
I still don't see the connection to LabVIEW.
Does Fluke have a support forum?
Looks like a hardware problem or maybe incorrect termination.
(594 Views)