9213 not returning actual temperatur values
I am using the NI USB 9162 carrier with a NI 9213. I need to read the temperature of some J type thermocouples and graph them for a period of 4 hours. After I start MAX I use Labview Signal Express. The step setup seems easy enough, but the values the thermocouple are returning seem off. I start it and it goes to about 1.376k in Celsius or 2. something in fahrenheit. I have applied heat with a hot air gun and let the thermocoule sit in ice, but it constantly shows the same value. Even when I try to calibrate it the numbers don't vary much.
I have a thernoneter with me that I connected the thermocouple to and it reads all the changes in temperature. Please help, I have tried everyhting I can think of.
I am brainstorming other options, like misplaced thermocouple connections inside the 9213 module, or my USB cable is too long to send a proper signal, it is only 6 feet though.
Please let me know if anyone can think of anything.
(2,591 Views)
Hi ceskyhlas,
It sounds like you may have an open thermocouple across your input.  Each channel has an open thermocouple detection (OTD) circuit, which consists of a current source between the TC+ and TC– terminals. If an open thermocouple is connected to the channel, the current source forces a full-scale voltage across the terminals.
Here are a few suggestions to help narrow in on the problem.
1. Perform a self test on the 9213 in MAX and ensure it passes.
2. Use test panels in MAX to test your 9213 to see if you can read a voltage level on your thermocouple.  At 25 degrees C you should get a voltage of 1.277 mV on a J type thermocouple.
3. See if you can independently measure a voltage on your thermocouple using a DMM.
4. Ensure that your leads on your thermocouple are connected properly- often times the red wire on the thermocouple is the (-) terminal.
5. Do you see the same readings on different channels?
Please let me know if you run into any problems after trying the above troubleshooting steps.
Thanks!
Brandon G
Precision DC Hardware Engineer
(2,573 Views)