Advise about DAQ..experts please get involved
Without amplification that might be hard. Looking at 16-bit AI you get a resolution of about 70 microV, with a 18-bit solution you'll get 0.8 microvolt resolution which should be enough (and sample speed should be enough with 625 kS/s). 18-bit resolution cost abit more, so scale the signal or get more sensitive gear, basically.
https://www.ni.com/en-us/support/model.pci-6280.html
That's the first example i could find, browse around, there might be other solutions.
/Y
Qestit Systems
Certified-LabVIEW-Developer
Message 11 of 15
(618 Views)
 Hi all
Yes there is amplification stage of course . but i meant to say that the electrical signal taken from the patient through gel electrodes is in the range of 70 microvolts . after that there is 3 amplification stages that turn the signl from microvolts range to millivolt. ( 1 mV approximately )
Tlaking about the sampling frequency .. well i guess that when programing m gonna set sampling frequency to be from 1000 samples to 4000 samples depending on the results.
Thank you so much !!
Bill David
Message 12 of 15
(613 Views)
Yamaeda.  That's not quite right.  The Minimum range accuracy for an 18 bit is probably going to be the driving factor.
Now let us discuss sample rate.  This is a bio sensor  human bodies are just not that fast! a 512 Hz sample rate is twice the speed of most state of the art implantable med devices Smiley Surprised Anything faster than that is a complete waste of processor since nothing that fast is affecting cell chemistry.  slow it down to keep the data meaningful.
Message 13 of 15
(610 Views)
Hi
okay . i understand about the sampling frequency . thank you so much . okay . do you advise me of NI DAQ6008 or 6009 ??
Bill David
Message 14 of 15
(606 Views)
with the 6008 in a -1 to +1 volt range your accuracy will be about 37.5mV with about 0.4mV resolution.  That doesn't sound ideal for a 1mV signal
Message 15 of 15
(598 Views)