Agilent 33220A error in burst mode

Hello everone,

 

I am running into the following error running Agilent 33220A in burst mode. I am suprirised the VI is working well until input frequency is 6 MHz, If I go beyond 6 MHz, I am getting this error.

Please help. My VI is also attahced

 

Agilent error-LV.png

 Agle

0 Kudos
Message 1 of 3
(2,514 Views)

Hi,

p 109 of user manual of the agilent " For sine and square waveforms, frequencies above 6 MHz are allowed only with an “infinite” burst count."

 

p 288 of same user manual. Error -221 "Settings conflict;frequency reduced for user function. For arbitrary waveforms, the output frequency is limited to 6 MHz. When you change to the arbitrary waveform function (APPL:USER or FUNC:USER command) from a function that allows a higher frequency, the function generator will automatically adjust the frequency to 6 MHz.

 

The problem does not come from your VI but from the limits of the generator

Yddet

Message 2 of 3
(2,466 Views)

Hi Yddet,

 

Thanks for the reply.

I came to know that.

 

 

0 Kudos
Message 3 of 3
(2,453 Views)