05-13-2020 02:35 PM
Hello everone,
I am running into the following error running Agilent 33220A in burst mode. I am suprirised the VI is working well until input frequency is 6 MHz, If I go beyond 6 MHz, I am getting this error.
Please help. My VI is also attahced
Agle
05-14-2020 04:46 AM
Hi,
p 109 of user manual of the agilent " For sine and square waveforms, frequencies above 6 MHz are allowed only with an “infinite” burst count."
p 288 of same user manual. Error -221 "Settings conflict;frequency reduced for user function. For arbitrary waveforms, the output frequency is limited to 6 MHz. When you change to the arbitrary waveform function (APPL:USER or FUNC:USER command) from a function that allows a higher frequency, the function generator will automatically adjust the frequency to 6 MHz.
The problem does not come from your VI but from the limits of the generator
Yddet
05-14-2020 11:48 AM
Hi Yddet,
Thanks for the reply.
I came to know that.