@Option13 wrote:

Oops, guess I uploaded the wrong version! I've been trying to hook it up to both RTDs and thermocouples and have configured the VI with both Thermocouple and RTD Vis. No change. The other VIs are just there for experimentation and to play with, if they present an issue I can remove them.


You are getting an initialization error.  Are you sure your VISA session is correct?


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 11 of 22
(2,039 Views)

I'll go back and check it, but it should be correct. I'm not currently hooked up to the device and ran the VI just to make sure the settings were correct and transferred to the TC sub VI. That's probably the source of that error. The Agilent takes data most of the day, so I can only connect at certain times because the Agilent can only support one session at a time. Makes this job a bit more difficult...

0 Kudos
Message 12 of 22
(2,033 Views)

@Option13 wrote:

Hello all,

I am attempting to get readings off of a thermocouple using an Agilent 34980A unit connected to my laptop via ethernet cable. The VI is based off of an Agilent provided example. It is connected to a T-type thermocouple plugged into channel 9, card 1. I have been able to get numbers and no error codes, however, the output values are basically either infinity or zero. 

 


That is often indicative of an open/bad thermocouple.

 

Does the thermocouple read properly if you program the instrument manually and read the display?

 

========================
=== Engineer Ambiguously ===
========================
0 Kudos
Message 13 of 22
(2,024 Views)

On an alternate, bespoke DAS the thermocouple reads just fine. After hooking up to the Agilent I changed the channel to see if that was the issue. Attached are the results screens. The error codes are why I changed to an RTD sensor, to see if those would read.

"1001" should be on card 1 (Agilent 34925A), channel 1, labeled "XS28" on our unit although I do not know if that is just on this unit. 

0 Kudos
Message 14 of 22
(2,018 Views)

Let's try to reduce where the error is comming from.  Do a scan of just the thermocoupler.  Do not configure or scan the RTDs.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 15 of 22
(2,013 Views)

Okay, I took it down to just the one thermocouple scan. The front panel shows 2.9118 E+01 (29.811 degrees Fahrenheit) and the same error codes.  

0 Kudos
Message 16 of 22
(2,009 Views)

I changed the reference junction. While I no longer get the error code, the output is still not correct. Setting to a fixed reference junction outputs +2.19272000E+01, and setting to external results in an output of -9.90000000E+37.

 

Again, the unit is collecting good data. How can I get LabVIEW to interpret it properly?

0 Kudos
Message 17 of 22
(1,982 Views)
Why do you say the unit is collecting good data? I know for a fact that the result displayed on the front panel was exactly what the program displayed.
0 Kudos
Message 18 of 22
(1,980 Views)

Because the Agilent is hooked up to a different data acquisition system when I am not trying to make it work with LabVIEW. I have not compared the LabVIEW readings it to the instrument front panel readings.

0 Kudos
Message 19 of 22
(1,975 Views)
Please make that comparison. You can manually configure the instrument, do a single scan, then programmatically fetch that reading.
0 Kudos
Message 20 of 22
(1,971 Views)