Agilent E4440A Display line

i try to set my Display Line at 6.299mW

 

i follow the manual attached, page 88's command.

 

It did turn on my Display Line, 

But the display line doesnt set to the value i want.

 

is it i write the command wrong ? :< 

 

help 

 

Noob_intern_0-1644457389698.png

Noob_intern_1-1644457414358.png

 

0 Kudos
Message 1 of 4
(1,230 Views)

Hi,

 

Manual says that:

" The line has an amplitude value that corresponds to its vertical position relative to the reference level."

Try to set it and check.

 

BR,

Aniket

0 Kudos
Message 2 of 4
(1,224 Views)

Hi Aniket

 

Appreciate that you replied.

 

I dont really understand the meaning of the sentence. 

Does it mean ,

for example if i want to get display line at 10 mW, and my reference line is 5mW, i need put 15 mW in the command ?

 

so 15 mW - 5 mW = 10 mW ?😥

 

Sorry if i understood wrongly, appreciate if you could explain or correct me :<

 

Thanks 

Intern w 1 month labview experience

0 Kudos
Message 3 of 4
(1,192 Views)

Yes, Or Simply set reference to 0